CT:
conventional CT (aka diagnostic CT)
Diagnosing
and
evaluating
treatment response
simulation CT
for radiation therapy treatment
planning
provide data on tissue
electron densities
HUs = pixels of differing shades of
grey
obtains dose
distribution
data
well defined anatomy is good for image
matching
image quality:
conventional
CT =
simulation
CT
Conventional and simulation CT:
fan
beam
1D
detector
1
rotation =
1
slice of CT
requires
multiple
rotations around patient
unlimited
scan length
HU =
yes
EPI: electronic portal imaging
MV
- opposing
inline
imager
in
compton
region of energy absorption - images have
low
contrast
BEV
images - captures true record of treatment
OBI: on board imaging
KV
-
perpendicular
to treatment
head
photoelectric
region of energy absorption - images have
high
contrast between tissues
used for
reference
images
used for
CBCT
- improved image
quality
CBCT - KV (via OBI)
cone
beam (instead of fan beam)
2D
detector
1 rotation =
entire
volume
requires
single
rotation around patient
scan length limited by detector
size
HU =
no
(so itβs not used for planning)
Orthogonal pair:
MV
-
KV
especially used when you need to obtain images
quickly
(ex. Patient is in pain)
Image types in RT:
DRR =
digitally reconstructed
radiograph
treatment
image
reference
image
BEV
= beams eye view
DRR:
planar
image - from
volumetric
CT simulation scan
provide
2D
representation of patient position in CT sim
use to compare
2D
daily treatment images to
3D
CT sim scan
can be generated for
reference
images and
BEVs
static
or
live
Static DRR:
generated in
external
beam planning
image is a
fixed
view
no
6DOF
couch to work with
Live DRR:
generated in
external
beam planning
updated dynamically by treatment software to account for changes in
pitch,
roll,
yaw
of patient
used when couch has
6DOF
Treatment image:
image taken on treatment unit (Linac) of the patient in treatment
position
can be:
reference image
,
BEV
,
CBCT
Reference image:
image taken with
collimator
jaws and
MLC
retracted to give full view of patient's
anatomy
used to compare patient
treatment
position to the
CT
sim (
planned
) position and for
localizing
the treatment fields
BEV:
image taken at treatment angle with
collimated
field size and
MLC
in place
visualization
of treatment field in relation to the patient's
anatomy
(is the field in the right location)
only used to confirm field
placements
not used for
image matching
(
field localization
)
Treatment image types:
2D
2D
/
2D
3D
2D
/
3D
2D image:
planar
image
provide
information
in
2
dimensions only
AP:
sup
/
inf
and
right
/
left
planes
lateral:
sup
/ inf and
ant
/
post
planes
2D / 2D: "
orthogs
"
two
planar
images assessed against
static
DRRs to give treatment field positioning information in all
3
planes
only allow for
translational
treatment field shifts (x, y, z)
two images taken at
90
degrees apart
3D:
volumetric
images - provide
3
dimensional data of a patient
ex.
CT
,
CBCT
,
MRI
2D / 3D:
same as
2D
/
2D
except treatment images are compared to
live DRRs
allows for
translational
and
rotational
shifts (x, y, z, pitch, roll, yaw)
used for
6DoF
treatment couches which can perform
rotational
shifts
Unpaired images:
two
orthogonal treatment images that share a
common
axis (sup / inf) but are
not linked
in the software
sup / inf shifts applied to one image will
NOT
be reflected in the other
Paired images:
two
orthogonal treatment images that share a
common
axis (sup / inf) that are
linked
together in the software
sup / inf shifts applied on one image
are
applied to the other
Summary of RT imaging:
A)
EPI
B)
BEV
C)
BEV
D)
MV
E)
KV
F)
MVCT
G)
CBCT
H)
CT
8
BEV exposure options:
Single Exposure:
closed
Double Exposure:
two
exposures on
one
image
one exposure =
closed
image (of the treatment
field
)
other exposure =
open
image (
larger
field to visualize surrounding
anatomy
)
BEV:
closed
image taken with
collimator jaws
and
MLC
in treatment
position
visualize
borders
of the treatment field
open
image taken with MLC
retracted
and collimator jaws
opened
wider
than
planned
field size to visualize anatomy
outside
of treatment field
Image only double:
image with
two
exposures (
open
and
closed
) taken either
before
or
after
the treatment
Before during:
image with
open
field taken before treatment and
closed
field taken during treatment
During after:
image with
closed
field taken during treatment
open
field taken after treatment
MRI: soft tissue contrast
T1 - weighted:
CSF
=
dark
,
fat
=
bright
T2 - weighted:
CSF
+
fat
=
bright
T1 with IV gadolinium contrast:
gadolinium = very
bright
on
T1
images
used for looking at
vascular
structures and
inflammation
FLAIR =
fluid suppression
suppressing fluid - help with visualization of abnormalities (
inflammation
)
similar to
T2
weighted - except:
TE
and
TR
are very
long
CSF =
dark
abnormalities =
bright
MRI sequences summary:
A)
dark
B)
bright
C)
dark
D)
bright
E)
dark
F)
bright
6
Fat Sat and STIR:
suppressing
fat
on MRI - improve visualization of pathologies or contrast
fat has short
T1
methods to suppress fatty tissues:
chemical shift
techniques (fat sat)
inversion recovery
(STIR)
hybrid of
chemical shift
and
IR
Diffusion weighted imaging:
provides information regarding
diffusion
properties of tissues
pathologies can affect
water
distributions inter and
intracellularly
used for
acute
strokes - can detect
ischemic
tissues