  • Accuracy
    Nearness of measured value to that of the target value
  • Diagnostic sensitivity

    The ability of a method to detect a population of individuals having the presence of disease
  • Standard
    A colorless solution with known concentration of substances used for calibration
  • Precision
    Nearness of measured values to each other
  • Quality
    A feature/characteristic of a product which meets the expected criteria of a consumer (customer)
  • Specificity

    The ability of a method to measure the analyte of interest only
  • Mean
    Average of a set of values (mean = Σx/n). Measures central tendency
  • Standard Deviation
    The measure of dispersion of values to that of the mean. Most frequently used measure of variation
  • Variance
    Square of SD. V=SD^2
  • Median
    Midpoint of a set of values
  • Range
    Simplest expression of spread or distribution
  • Intralab QC (internal QC)

    Control samples are run simultaneously with a patient to ensure reliability of methods and result. Used for daily monitoring of accuracy and precision of method used. Detects random and systematic errors
  • Delta check is used to check if there are significant differences between present set of values to that of past values on the sample of the same individual
  • Systematic errors are tested for inaccuracy using 22s, 41s, and 10x. Sources of systematic errors include improper calibration, reagent deterioration, contaminated solutions, instability of both samples and solutions
  • Shewhart Levey-Jennings Chart

    Most widely used QC chart
  • Trend
    Six or more consecutive values that either increase or decrease gradually (will cross the mean). Main cause: reagent deterioration
  • Six Sigma is a way of improving product processing to eliminate defects
  • Wavelength
    The distance between two successive peaks
  • Nanometer
    Unit expression of wavelength
    1. test
    Used to assess if there is a statistical difference between the means of 2 groups of data
  • Diagnostic specificity

    The ability of a method to detect a population of individuals absent of a disease process
  • Sensitivity
    The ability of a method to measure analytes even at its lowest concentration
  • Control

    A solution that resembles a human sample that is used for QC purposes only
  • Coefficient of variation
    Mean expression in percentile. Index of precision
  • Mode
    The most frequent among all values/data
  • Interlab QC (external QC)
    Laboratories are given samples with unknown concentrations for them to test and results are compared with other laboratories thus maintaining “long-term accuracy” to methods utilized. Results difference of greater than 2SD indicates disagreement with other lab included
    1. test
    Used to assess if there is a statistical difference between the SD of 2 groups of data
  • Random errors are tested for imprecision using 12s (warning rule), 13s, and R4s. Sources of random errors include mislabeling, pipetting error, fluctuations in temperature & voltage
  • Energy is an entity that is transmitted by electromagnetic waves
  • Shift
    Six or more consecutive values that are distributed on one side or other side of the mean (does NOT cross the mean). Main cause: improper instrument calibration
  • Beer-Lambert Law
    States that concentration of an unknown analyte is directly proportional to the light absorbed and inversely proportional to light transmitted
  • Double-Beam Spectrophotometer
  • Flame Emission Photometry

    1. Flame permits the excitation of electrons; radiation is emitted when electrons return to a ground state
    2. Internal standards used: Cesium and Lithium
  • Wavelength
    Distance between two successive peaks
  • Frequency
    Number of waves that pass a point of observation per one unit of time
  • Parts of Spectrophotometer
    • Light Source
    • Entrance Slit
    • Monochromator
    • Exit Slit
    • Cuvette
    • Photodetector
    • Read-Out Device
  • Fluorometry
    1. Light is absorbed at a specific wavelength and emitted at a longer wavelength
    2. Light source: xenon lamp or mercury arc
    3. Two monochromators: Primary monochromator selects wavelength best absorbed by solution
  • Definition of Terms in Clinical Chemistry:
  • Beer-Lambert Law

    Absorbance is proportional to the inverse log of transmittance
  • Spectrophotometry
    1. Measures transmitted light in a colored solution
    2. Based on Beer-Lambert-Bouguer Law (Beer’s Law/Beer-Lambert’s Law)