analytical spectroscopy

    Cards (68)

    • most IR and NMR spectroscopy is
      qualitative
    • uv-vis spectroscopy is
      quantitative
    • qualitative analysis
      analyses are identified on the basis of the wavelength(s) that they absorb or emit
    • quantitative analysis
      concentration of analytes are measured
    • selectivity
      degree to which a method responds to the analyte and not to other components of the sample matrix
    • selectivity must be
      high so the method is free from interference
    • sensitivity
      change in response per unit change in concentration of the analyte
    • sensitivity must be
      high so we can distinguish small differences in concentration of the analyte
    • limit of detection
      lowest analyte concentration we can detect with a particular statistical certainty
    • what is uv-vis spectroscopy
      light in the wavelength of 200-800 nm can induce electronic transitions (movement of valence electrons between orbitals)
      so atoms or molecules can absorb or emit uv-vis light
      the wavelength of light absorbed allows compounds to be identified
      the amount of light absorbed relates to concentration
    • describe qualitative uv-vis spectroscopy
      the wavelength at which a substance shows maximum absorption is called the (lambda max) corresponds to the delta E of the transition occurring
      can be used to id unknowns since each substance has a different lambda max
    • LAMBDA MAX does not correspond to the colour of the solution
      we see the un-absorbed portion of the visible spectrum, the complementary colour
    • how is quantitative analysis preformed
      light of correct wavelength (mathching delta E of the analyte) is shone through the solution
      the proportion of light absorbed is measured
    • incident radiation (I0) is ...
      100
    • I (transmitted radiation)
      is x
    • Absorbance (A)
      log10(I00/I)
    • the beer lambert law relates
      absorption to concentration
    • the beer lambert law is 

      A = e c l
    • e
      molar absorptivity dm3 mol-1 cm -1
    • c
      concentration mol dm-3
    • l
      path length (cm)
    • molar absorptivity relates to the probability of a particular electronic transition occurring.
    • large e values correspond to
      strong and intense colours
    • small e values correspond to
      pale and weak colours
    • large e values give good sensitivity in spectroscopic analysis and so are desirable
    • a chromophore is the specific functional group in the molecule that absorbed a particular wavelength of light
    • different chromophores have different, characteristic, absorption wavelengths and e values
    • common chromophores in uv-vis spectrometry include
      double or triple bonds and conjugated systems
      lone pairs
      metal ion complexes
    • double or triple bonds, and conjugated systems such as benzene or butadiene have e values in the range of
      1000 - 100000 dm3 mol-1 cm-1
    • lone pairs e.g. on O, N or S atoms have e values around
      1000 dm3 mol-1 cm-1
    • metal ion complexes have e values in the range of
      0.3 - 100000 dm3 mol-1 cm-1
    • examples of organic chromophores include
      amines, alkenes, conjugated alkenes, ketones, aromatics
    • direct spectrophotometry finds the concentration of the unknown with respect to a series of standard solutions made up in the same solvent and measured at the same temperature
    • direct spectrophotometry overcomes the fact that molar absorbivity is not a universal constant but instead varies because of temperature and the composition of the solvent the analyte is dissolved in which can make using the beer-lambert law to accurate difficult
    • in DS
      1 - the wavelength which gives maximum absorbance for the analyte of the interest is found
      2 - set the spectrometer and determine the analyte concentration in the sample by
      making up a series of standard solution of known concentrations of the analyte and one blank
      analyse the standards
      plot a calibration graph
      analyse the samples
      use the calibration graph to obtain the concentration of analyte in the samples
    • limitation of DS include
      only linear part of the graph is used (usually up to A = around 0.7)
      the response for the sample must lie within the linear portion of the graph
      new graph for each analysis
    • light source (DS) - a tungsten and/or a deuterium lamp to provide visible and uv wavelengths, respectively
    • monochromator (DS) - selects a small wavelength region from the lamps output spectrum to pass through the sample (typical bandpass is 2nm)
    • cell (DS) - contains the sample solution
    • detector (DS) - a photomultiplier which gives out an electrical signal when light falls upon it