analytical spectroscopy

Cards (68)

  • most IR and NMR spectroscopy is
    qualitative
  • uv-vis spectroscopy is
    quantitative
  • qualitative analysis
    analyses are identified on the basis of the wavelength(s) that they absorb or emit
  • quantitative analysis
    concentration of analytes are measured
  • selectivity
    degree to which a method responds to the analyte and not to other components of the sample matrix
  • selectivity must be
    high so the method is free from interference
  • sensitivity
    change in response per unit change in concentration of the analyte
  • sensitivity must be
    high so we can distinguish small differences in concentration of the analyte
  • limit of detection
    lowest analyte concentration we can detect with a particular statistical certainty
  • what is uv-vis spectroscopy
    light in the wavelength of 200-800 nm can induce electronic transitions (movement of valence electrons between orbitals)
    so atoms or molecules can absorb or emit uv-vis light
    the wavelength of light absorbed allows compounds to be identified
    the amount of light absorbed relates to concentration
  • describe qualitative uv-vis spectroscopy
    the wavelength at which a substance shows maximum absorption is called the (lambda max) corresponds to the delta E of the transition occurring
    can be used to id unknowns since each substance has a different lambda max
  • LAMBDA MAX does not correspond to the colour of the solution
    we see the un-absorbed portion of the visible spectrum, the complementary colour
  • how is quantitative analysis preformed
    light of correct wavelength (mathching delta E of the analyte) is shone through the solution
    the proportion of light absorbed is measured
  • incident radiation (I0) is ...
    100
  • I (transmitted radiation)
    is x
  • Absorbance (A)
    log10(I00/I)
  • the beer lambert law relates
    absorption to concentration
  • the beer lambert law is 

    A = e c l
  • e
    molar absorptivity dm3 mol-1 cm -1
  • c
    concentration mol dm-3
  • l
    path length (cm)
  • molar absorptivity relates to the probability of a particular electronic transition occurring.
  • large e values correspond to
    strong and intense colours
  • small e values correspond to
    pale and weak colours
  • large e values give good sensitivity in spectroscopic analysis and so are desirable
  • a chromophore is the specific functional group in the molecule that absorbed a particular wavelength of light
  • different chromophores have different, characteristic, absorption wavelengths and e values
  • common chromophores in uv-vis spectrometry include
    double or triple bonds and conjugated systems
    lone pairs
    metal ion complexes
  • double or triple bonds, and conjugated systems such as benzene or butadiene have e values in the range of
    1000 - 100000 dm3 mol-1 cm-1
  • lone pairs e.g. on O, N or S atoms have e values around
    1000 dm3 mol-1 cm-1
  • metal ion complexes have e values in the range of
    0.3 - 100000 dm3 mol-1 cm-1
  • examples of organic chromophores include
    amines, alkenes, conjugated alkenes, ketones, aromatics
  • direct spectrophotometry finds the concentration of the unknown with respect to a series of standard solutions made up in the same solvent and measured at the same temperature
  • direct spectrophotometry overcomes the fact that molar absorbivity is not a universal constant but instead varies because of temperature and the composition of the solvent the analyte is dissolved in which can make using the beer-lambert law to accurate difficult
  • in DS
    1 - the wavelength which gives maximum absorbance for the analyte of the interest is found
    2 - set the spectrometer and determine the analyte concentration in the sample by
    making up a series of standard solution of known concentrations of the analyte and one blank
    analyse the standards
    plot a calibration graph
    analyse the samples
    use the calibration graph to obtain the concentration of analyte in the samples
  • limitation of DS include
    only linear part of the graph is used (usually up to A = around 0.7)
    the response for the sample must lie within the linear portion of the graph
    new graph for each analysis
  • light source (DS) - a tungsten and/or a deuterium lamp to provide visible and uv wavelengths, respectively
  • monochromator (DS) - selects a small wavelength region from the lamps output spectrum to pass through the sample (typical bandpass is 2nm)
  • cell (DS) - contains the sample solution
  • detector (DS) - a photomultiplier which gives out an electrical signal when light falls upon it