advanced imaging

Cards (29)

  • Computed tomography (CT)

    Advanced imaging technique that uses X-rays to create cross-sectional images of the body
  • Main CT components
    1. CT gantry (tube, filters, detectors) and patient table
    2. Computer
    3. CT operating console and workstation
    PACS
  • The CT scanning room
    • Gantry
    Patient table
    Shielding
    Room for stretchers, resuscitation team, emergency equipment, medical gases
    Injector for IV contrast
    Control room with computers, consoles and workstations
  • CT table
    Motorised platform which patient lies on for CT examination
    Lightweight (carbon)
    Differs between diagnostic radiology (DR) and radiotherapy (RT) machines
    DR tables dipped for patient comfort, RT tables flat to ensure reproducibility
  • CT gantry
    Houses the data acquisition system
    Rotating frame has X-ray tube mounted on one side and detectors on the other
    Controls to lift/move table
    Tilt range
    Laser lights for centring/localisation
    RT: wider bore
  • Inside the CT gantry
    1. ray tube
    2. 4. Collimators
    5. Detectors
  • CT scanner design
    • X-ray tube needs to withstand higher energies, currents and scan times compared to standard X-ray
    Bow-tie filter filters beam to reduce skin dose
    Collimators restrict X-ray beam to detector array and reduce scatter radiation
    Scintillation detectors efficiently convert X-ray photons to electrical signal
  • Slip ring technology
    Allows continuous rotation of the gantry and table feed, enabling faster scanning
  • Helical scanning
    Images acquired constantly as table moves slowly through gantry
    Faster scans, reduced movement artefacts
    But susceptible to cone beam artefact and requires z-interpolation
  • Sequential/axial scanning

    Stop-start rotation of the gantry, slower scans
  • Multi-slice CT

    Multiple detector rows in the z-axis, enabling multiple slices per rotation
    Faster acquisition, reduced movement artefacts, longer scan volumes, good for CT angiography
  • Benefits of multi-slice CT
    • Allows sub-millimeter scanning, multiplanar reformats, 3D reconstructions, high resolution CT
    Ability to choose thicker or thinner slice thickness for different purposes
  • Anisotropic resolution
    Voxel dimensions in x, y and z planes are not similar, leading to 'stair-step' artefacts
  • Isotropic resolution
    Voxel dimensions in x, y and z planes are similar, providing greater spatial and contrast resolution
  • CT scanning indications
  • Patient preparation
    Pregnancy check, contraindications to contrast agent, changed into gown
    Intravenous contrast to visualise blood vessels and organ vasculature
    Oral contrast to visualise GI tract
  • Patient positioning
    Patient must be centred into the isocentre of the gantry using laser lights
    Centring in isocentre reduces dose, optimises image quality and reduces artefacts
  • Scout scan
    Allows planning of slices and field of view, checks for artefacts, allows calculation of mA modulation
  • mA modulation
    Varies the tube current along the patient to optimise image quality
  • Pitch
    Ratio of table movement per rotation to slice thickness
    Increasing pitch decreases scan time but reduces scan quality, decreasing pitch increases scan time but improves quality
  • Field of view (FOV)
    Radiotherapy - wide FOV to see full outline of patient
    Diagnostics - variable FOV depending on area of interest
    Scan FOV vs display FOV
  • Intravenous contrast agent
    Injected by pump, scan activated at specific time intervals to visualise arterial or portal venous phases
  • Bolus tracking
    Used when looking at arterial system, scan triggered when contrast reaches a certain density in the vessel of interest
  • During the CT scan
    No one except patient in room, watch patient, check for contrast reaction, explain waiting for image reconstruction
  • Viewing CT images
    Check images immediately post-acquisition, look for artefacts, pathology outside field of view, incidental findings
  • Common CT artefacts
    • Beam hardening
    Motion
    Ring artefact
    Partial volume artefact
  • CT dose

    CTDIvol - absorbed dose per unit mass of irradiated volume
    DLP - dose length product, indicative of total radiation to patient
    National diagnostic reference levels (DRLs) to standardise doses
  • Image post-processing
    Reformats, magnification, window adjustments, multiplanar reconstructions, maximum intensity projections, volume rendering
  • Methods to optimise dose and improve image quality
    • Optimisation of protocols
    Limit scanning to only necessary area
    Staff training
    Daily QA and air calibration of scanner
    Dose reduction software (mA modulation, iterative reconstruction)
    Use appropriate protocol (e.g. low dose, paediatric)