Must To Know in Clinical Chemistry

Cards (738)

  • Quality Control
    • Practicability
    • Reliability
    • Maintain accuracy and precision
    • Intralab/Interlab QC
    • Daily monitoring of accuracy and precision
    • Interlab/External QC
    • Proficiency testing (Reference lab)
    • Long-term accuracy
  • Difference of >2
    Not in agreement with other lab
  • QC materials
    • Available for a min. of 1 yr
    • Bovine control materials (preferred, human biohazard)
    • Not for immunochem, dye-binding and bilirubin
  • Matrix effect
    • Improper product manufacturing
    • Unpurified analyte
    • Altered protein
  • Precision study
    First step in method evaluation
  • Nonlab. personnel
    29% of errors (lab results)
  • SD
    Dispersion of values from the mean
  • CV
    • Index of precision
    • Relative magnitude of variability (%)
  • Variance
    • SD2
    • Measure of variability
  • Inferential statistics

    Compare means or SD of 2 groups of data
    1. test
    Means of 2 groups of data
    1. test
    SD of 2 groups of data
  • Cumulative Sum Graph (CUSUM)

    Earliest indication of systematic errors (trend)
  • Youden/Twin Plot

    Compare results obtained from different labs
  • Shewhart Levey-Jennings Chart

    Graphic representation of the acceptable limits of variation
  • Trend
    • Gradual loss of reliability
    • Cause: Deterioration of reagents (Systematic error)
  • Shift
    • Values: one side or either side of the mean
    • Cause: Improper calibration (Systematic error)
  • Outliers
    • Values: far from the main set of values
    • Highly deviating values
    • Random or systematic errors
  • Kurtosis
    Degree of flatness or sharpness
  • Precision
    Random error
  • Accuracy
    Systematic error
  • Causes of Random error
    • Mislabeling
    • Pipetting error
    • Improper mixing of sample and reagents
    • Voltage/Temperature fluctuation
    • Dirty optics
  • Causes of Systematic error
    • Improper calibration
    • Deterioration of reagents
    • Contaminated solution
    • Sample instability/unstable reagent blanks
    • Diminishing lamp power
    • Incorrect sample and reagent volume
  • Parameters for Random error
    SD and CV
  • Multirule Shewhart procedure
    Control rules + Control chart
  • Test method
    Westgard: at least 40 samples
  • Reference method
    Westgard: preferably 100 samples
  • Analytical Run
    Control and patient specimens assayed, evaluated, and report together
  • Physiologic Limit
    Referred to as absurd value
  • POCT
    Performed by nonlab personnel
  • Quality Assurance
    • Program development
    • Assessment and monitoring
    • Quality improvement
  • Quality Patient Care
    Test request forms, clear instruction for patient prep., specimen handling...
  • Reference Range/ Interval
    At least 120 individuals should be tested in each age and sex category
  • Wavelength
    • Distance between 2 successive peaks (nm)
    • Lower frequency = Longer wavelength (Ex. Red)
    • Higher frequency = Shorter wavelength (Ex. Violet)
  • Spectrophotometric measurement
    Measures light intensity in a narrower wavelength
  • Photometric measurement

    Measures light intensity without consideration of wavelength
  • Multiple wavelength

    Uses filter only
  • LASER
    • Light Amplification by Stimulated Emission of Radiation
    • Light source for spectrophotometry
  • Light sources
    • Visible region: Tungsten light bulb, Mercury arc
    • UV: Deuterium lamp, Mercury arc, Xenon lamp, Hydrogen lamp
    • IR: Merst glower, Globar (Silicone carbide)
  • Stray light
    • Wavelength outside the band
    • Most common cause of loss of linearity