For any given radiographic examination, the numberofx-rays that reach the image receptor is directly related to both the x-ray tube current and the time that the x-ray tube is energized
Exposure timers
1. X-ray operating consoles provide a wide selection of x-ray beam-on times
2. Used in conjunction with the appropriate mA station to provide an even wider selection of values for mAs
Guard timer
A timing circuit that will terminate an exposure after a prescribed time, usually approximately 6 s
Timer circuit
1. Makes and breaks the high voltage across the x-ray tube
2. Nearly always done on the primary side of the high-voltage transformer, where the voltage is lower
Types of timing circuits
Synchronous timers
Electronic timers
mAs timers
Automatic exposure control
Synchronous timers
Driven by a synchronous motor to provide precise 60 rps timing
Minimum exposure time is 1/60 s (17 ms)
Cannot be used for serial exposures as they must be reset after each exposure
Electronic timers
Most sophisticated, complicated, and accurate
Based on the time required to charge a capacitor through a variable resistance
Allow a wide range of time intervals to be selected, accurate to 1 ms
Suitable for rapid serial exposures
mAs timers
Monitor the product of mA and exposure time and terminate exposure when the desired mAs value is attained
Provide the highest safe tube current for the shortest exposure for any mAs selected
Located on the secondary side of the high-voltage transformer
Automatic exposure control (AEC)
Measures the quantity of radiation that reaches the image receptor
Automatically terminates the exposure when the image receptor has received the required radiation intensity
Uses an ionization chamber or photodiode detector assembly
AEC calibration
1. Making exposures of a test object and adjusting the AEC for the required x-ray intensities
2. Usually done by the service engineer
AEC mode
Requires particular care, especially in low kVp examinations like mammography
The AEC may not respond properly due to varying tissue thickness and composition
When using AEC
The electronic timer should be set to 1.5 times the expected exposure time as a backup in case the AEC fails to terminate
High-voltage generator
Responsible for increasing the output voltage from the autotransformer to the kVp necessary for x-ray production
High-voltage transformer
A step-up transformer with a turns ratio usually between 500:1 and 1000:1
Operates only on alternating current, producing a sinusoidal voltage waveform
Turns ratio of high-voltage transformer
Voltage increase is proportional to the turns ratio, according to the transformer law
Voltage rectification
Converting alternating current (AC) to direct current (DC) to provide the x-ray tube with electron flow in only one direction
Rectifier
An electronic device that allows current flow in only one direction
rays are produced by the acceleration of electrons from the cathode to the anode and cannot be produced by electrons flowing in the reverse direction</b>
Semiconductor
A material that lies between the range of insulators and conductors in the ability to conduct electricity
Semiconductors are classed into two types: n-type and p-type
Voltage rectification
Rectifiers in most modern x-ray generators are the silicon, semiconductor type
Semiconductors
Lie between the range of insulators and conductors in the ability to conduct electricity
Tiny crystals have some useful electrical properties and allow semiconductors to serve as the basis for today's solid-state microprocessor marvels
Types of semiconductors
N-type
P-type
type semiconductors
Have loosely bound electrons that are relatively free to move
type semiconductors
Have spaces, called holes, where there are noelectrons. Holes are as mobile as electrons.
n junction
1. Electrons and holes migrate toward the junction and wander across it
2. Flow of electrons and holes constitutes an electric current
Positive potential on n side of p-n junction
Electrons and holes swept away from junction, no current passes through
n junction
Conducts electricity in only one direction, called a solid-state diode
Rectifiers
Conduct electric current in only one direction
Rectification is essential for the safe and efficient operation of the x-ray tube
Unrectified voltage
Voltage waveform on secondary side of high-voltage transformer
Current through x-ray tube only exists during positive half of cycle when anode is positive and cathode is negative
Half-wave rectification
Voltage not allowed to swing negatively during negative half of cycle
Half-wave rectified circuit
1. Contains one or more diodes
2. Allows current to pass through x-ray tube during positive portion of AC waveform
3. Does not conduct during negative portion, resulting in a series of positivepulses
Full-wave rectification
Negative half-cycle reversed so anode is always positive
Full-wave rectified circuit
1. Contains at least fourdiodes
2. Current passed through tube at 120 pulses per second
Main advantage of full-waverectification is that exposure time for any given technique is cut in half
Full-waverectification
1. Electrons flow through diodeD and x-raytube during positivehalf-cycle
2. Electrons flow through diodeA during negativehalf-cycle
Half-wave rectified x-ray tube emits x-rays only half of the time, full-wave rectified emits 120 times per second
Single-phase power
Results in a pulsating x-ray beam due to alternate swing in voltage from zero to maximum 120 times per second