Cards (30)

  • Vaporizer
    An instrument used to vaporize volatile anesthetic agents (halothane, isoflurane, sevoflurane, desflurane) and deliver it to the patient as a controlled amount of an inhalational agent, after changing it from liquid to vapor, through the fresh gas flow
  • Vaporizer
    • Must be located between the flowmeter and common gas outlet
    • Normally expressed as a percentage of saturated vapor added to the gas flow like 1%, 2%, 0.75%
    • All anesthetic machines should have an interlocking or exclusion device to prevent concurrent use of more than one vaporizer
  • Vapor pressure
    The pressure exerted by vapor in equilibrium with solid or liquid of the same substance
  • As temperature increases
    The vapor pressure increases
  • The gas molecules bombard the walls of the container, creating the saturated vapor pressure of that agent
  • Vaporization
    Requires energy (the latent heat of vaporization), which results in a loss of heat from the liquid
  • As vaporization proceeds
    The temperature of the remaining liquid anesthetic drops and vapor pressure decreases unless heat is readily available to enter the system
  • Boiling point
    The temperature at which a liquid's vapor pressure is equal to the atmospheric pressure
  • As atmospheric pressure decreases (as in higher altitudes), the boiling point also decreases
  • Anesthetic agents with low boiling points are more susceptible to variations in barometric pressure than agents with higher boiling points
  • Among the commonly used agents, desflurane has the lowest boiling point (22.8°C at 760 mm Hg)
  • Characteristics of the ideal vaporizer
    • Its performance is not affected by changes in FGF, volume of the liquid agent, ambient temperature and pressure
    • Low resistance to flow
    • Light weight with small liquid requirement
    • Economy and safety in use with minimal servicing requirement
    • Corrosion- and solvent-resistant construction
  • Draw-over vaporizer

    Gases pass through a very low resistance vaporizer due to the patient's respiratory efforts
  • Plenum vaporizer
    Gases are driven through a high resistance, unidirectional and agent specific vaporizer due to gas supply pressure
  • Conventional (variable-bypass) vaporizers
    • The total background gas flow that enters the unit is split into two streams: a smaller stream passes through the vaporizing chamber and becomes saturated with agent vapor, the remainder of the gas bypasses this chamber
    • A wick may be used in the vaporizing chamber to provide increased surface area for evaporation
    • One adjustment is made to get desired concentration which simultaneously balances the carrier and bypass flows to produce the blend required for the set concentration
  • Components of a conventional vaporizer

    • The case with the filling level indicator and a port for the filling device
    • Percentage control dial on top of the case
    • The bypass channel & the vaporization chamber with or without wick
    • The splitting ratio is controlled by a temperature-sensitive valve utilizing a bimetallic strip
    • The vaporizers are mounted on the back bar using the interlocking Selecta Tec system
  • The FGF only enters the vaporizer when it is switched on
  • Cooling effect

    During vaporization, cooling occurs due to the loss of latent heat of vaporization. Lowering the temperature of the agent makes it less volatile.
  • Compensation for cooling effect

    • The vaporizer is made of a material with high density and high specific heat capacity with a very high thermal conductivity, e.g. copper
    • A temperature sensitive valve (e.g. bimetallic strip or bellows) within the body of the vaporizer automatically adjusts the splitting ratio according to the temperature
  • Ventilators exert back pressure as they cycle may back into the vaporizing chamber, where more vapor is added
  • Compensation for back pressure

    • Long inlet port into the vaporizing chamber
    • Downstream flow restrictors
    • Both the bypass channel and the vaporizing chamber are of equal volumes so gas expansion and compression are equal
  • A pressure relief valve downstream of the vaporizer opens at about 35 kPa to prevent damage to flowmeters or vaporizers if the common gas outlet is blocked
  • Heated-blender vaporizers

    Designed for desflurane (Suprane) due to its higher volatility compared to other agents
  • Heated-blender vaporizers

    • Maintain a constant desflurane vapor pressure and allow accurate delivery
  • Electronic vaporizers
    Electronically controlled vaporizers must be utilized for desflurane and are used for all volatile anesthetics in some sophisticated anesthesia machines
  • Desflurane vaporizer

    • Contains a reservoir of desflurane (desflurane sump) that is electrically heated to 39°C, creating a vapor pressure of 2 atmospheres
    • No fresh gas flows through the desflurane sump, pure desflurane vapor joins the fresh gas mixture before exiting the vaporizer
    • Decreased ambient pressure (high elevation) does not affect the concentration
  • Aladin cassette vaporizer
    A novel system that regulates anesthetic vapor concentration with electro-pneumatic proportional flow valve controlled by microprocessors
  • Aladin cassette vaporizer

    • Provides an accurate amount of anesthetic agent while mitigating the effect of temperature and barometric pressure on evaporation process
    • Automated recognition of inserted gases
    • On-screen data show agent level and usage
    • Automated electronically monitored controlled FGF, temperature, pressure compensation
    • No risk of spillage into bypass channel
    • Can be safely carried in any orientation
  • Vaporizer filling devices

    • Agent-specific, color coded to fit the safety filling port of the correct vaporizer and anesthetic agent supply bottle
    • Prevent the risk of adding the wrong agent to the wrong vaporizer
    • Decrease the extent of spillage
    • Ensure that the vaporizer cannot overflow
  • Anesthetic agents

    • Halothane
    • Isoflurane
    • Sevoflurane
    • Desflurane