Collects and discards waste anesthetic gases safely
Scavenging system
Should not affect the ventilationandoxygenation of the patient
Should not affect the dynamics of the breathing
Basic components of any type of scavenger
Collecting and transfer system
Receiving system
Disposal system
Open scavenging system
Uses a reservoir to collect waste gas, which is then actively suctioned from the reservoir. The reservoir communicates with the atmosphere and requires no pressure relieving valve.
Closed scavenging system
Not connected to the atmosphere of the OR but waste gas exits the machine and then the OR through a series of valves. Must have both positive- and negative-pressure relief valves.
Types of closed scavenging system
Active via means of suction
Passive, relying on heavier-than-air anesthesia gases and pressure to move waste gas to evacuation
Passive scavenging system
Simple
Low cost
Require positive and negative pressure relieving valve for protection from collapse and barotrauma
Outlet may be a direct line to the outside via a ventilation duct beyond any point of recirculation
Components of passive scavenging system
Collecting and transfer system
Receiving system (reservoir bag)
Disposal system (wide-bore copper pipe)
Mechanism of passive scavenging system
1. Exhaled gases are driven by either the patient's respiratory efforts or the ventilator
2. Receiving system should be mounted on the anesthetic machine to minimize the length of transfer tubing and resistance to flow
Problems in practice and safety features of passive scavenging system
Components of active scavenging system
Collecting and transfer system
Receiving system (valveless, open-ended reservoir)
Active disposal system (fan or pump to generate vacuum)
Mechanism of active scavenging system
1. Vacuum drives the gases through the system
2. Able to deal with a wide range of expiratory flow rates (30–130 L/min)
3. Receiving system is capable of coping with changes in gas flow rates
Problems in use of active scavenging system
Scavenging system checkout
Ensure proper connections
Adjust waste-gas vacuum (if possible)
Fully open APL valve and occlude Y-piece
With minimum O2 flow, allow scavenger reservoir bag to collapse completely and verify absorber pressure gauge reads about zero
With the O2 flush activated, allow scavenger reservoir bag to distend fully, and then verify absorber pressure gauge reads <10 cm H2O
Charcoal canisters (Cardiff Aldasorber)
Compact passive scavenging system that absorbs halogenated inhalational agents
Problems and safety features of charcoal canisters