Location – remote from traffic and potentially disturbing air currents
Cabinets should be turned on at least 5 min before beginning work and after completion of work to allow the cabinet to "purge"
Materials to be placed inside the cabinet should be surface decontaminated with 70% alcohol
Ensure the front intake grill of BSC is not blocked with paper, equipment or other items
All materials should be placed as far back in the cabinet, towards the rear edge of the work surface, without blocking the rear grill
Place all necessary items into the cabinet before beginning manipulations
Arms should be moved in and out of the BSC slowly, and manipulations within the BSC should be delayed for about 1 min after placing hands and arms inside to allow the cabinet to adjust and to "air sweep" the surface of the hands and arms
Aerosol-generating equipment (e.g. mixers, centrifuges, etc.) should be placed towards the rear of the cabinet
Observe "clean to dirty work flow/set-up"
Open flames should be avoided inside the BSC, as they disrupt the air flow patterns and can be dangerous when volatile, flammable substances are also used