Review 1.5

Cards (40)

  • Waveforms
    Graphical representation of ventilator parameters over time
  • Interpretation of waveforms has improved
    Recognition rate of asynchrony and its clinical outcomes
  • Ventilator graphics
    Allow assessment and evaluation of the adequacy of inspiratory flow during Volume ventilation
  • Ventilator graphics
    Displayed in scalar and loop forms
  • Scalar waveforms

    • Flow-time
    • Volume-time
    • Pressure-time
  • Loop waveforms
    • Flow-volume
    • Pressure-volume
  • Scalar
    Single variable displayed against time, main components are pressure, flow, volume
  • Time axis
    X axis
  • Flow, Volume, Pressure axis
    Y axis
  • Flow-time scalar
    • Inspiration above and expiration below baseline
  • Square waveform in VC mode
    Helps reduce severe air trapping, and for recruitment maneuvers
  • Decelerating waveform in VC mode
    Patients tend to tolerate better as it mimics natural breathing
  • Flow-time waveform in PCV mode
    Decelerating flow
  • Flow-time waveform in PSV mode
    Sine wave flow
  • Airway obstruction
    Prolonged expiratory flow and decrease in Peak Expiratory Flow Rate (PEFR)
  • Positive response to bronchodilator therapy
    Return to baseline in a shorter period and increase in PEFR
  • Air trapping
    Expiratory flow does not return to baseline, new breath starts below baseline
  • Air trapping
    Also known as Autopeep
  • Causes of air trapping
    Short expiratory time, long inspiratory time, high respiratory rate
  • Volume-Time Scalar
    Visual representation of inspiratory and expiratory tidal volume
  • Volume waveform
    Should begin and end at baseline
  • Air leaks
    Assessed by looking at Volume-Time Scalar
  • Causes of air leaks
    Volume loss through circuit, chest tubes, bronco-pulmonary fistula, decrease in cuff pressure
  • Pressure-Time Scalar

    Shows visual representation of PIP and PEEP
  • Baseline pressure

    Also known as PEEP level
  • Rise time
    Rate at which flow increases from baseline to peak during pressure-support or pressure-controlled breaths
  • Rapid rise time
    • Bump in the pressure-time waveform at beginning of inspiration
  • Slow rise time
    • Pressure-time waveform is more curved and less square
  • Pressure-Volume Loop direction
    Counter-clockwise for mechanical breath, clockwise for spontaneous breath
  • Pressure-Volume Loop shift to the right
    Compliance decreases
  • Pressure-Volume Loop shift to the left
    Compliance increases
  • Lung overdistention on Pressure-Volume Loop
    • Bird Beak, excessive stretching or inflation of the lung
  • Air leaks on Pressure-Volume Loop

    Expiratory waveform does not return to zero volume/set PEEP
  • Flow-Volume Loop
    Displays peak inspiratory flow rate, peak expiratory flow rate, tidal volume
  • Flow-Volume Loop
    • Can assess air leaks, increased airway resistance, and airway secretions/condensation accumulation
  • Air leak on Flow-Volume Loop
    • Loop never closes, gap indicates magnitude of leak
  • Increased airway resistance
    Scoop in expiratory tracing of flow-volume loop, decrease in PEFR
  • Increased airway resistance on Flow-Volume Loop

    • Scooped out pattern, universal sign for obstructive lung disease
  • Airway secretions or condensate on Flow-Volume Loop

    • Sawtooth pattern on inspiration or expiration
  • Know the PFT Flow Volume Loops of Respiratory Diseases