Review 1.3

Cards (19)

  • Flow asynchrony
    Inspiratory gas flow from the vent is unable to match the patients inspiratory flow demand
  • Flow asynchrony
    • Scoop waveform
  • VC/AC mode
    Ventilator mode where flow asynchrony is commonly seen
  • Increase the inspiratory flow

    Decrease in I time and increase in E time
  • Decrease the inspiratory flow

    Increase in I time and decrease in E time
  • For adults, peak flow values are generally kept between 60-80 L/min
  • For adults, the typical I-time is 0.8 to 1 second
  • Square waveform
    Lower peak flow and decreased inspiratory time, compared to decelerating
  • Decelerating waveform
    Mimics more natural breathing pattern, produces high initial peak flow but increased iT, pts are usually placed on this waveform on VC AC mode
  • Rise time
    Allows adjustment of the rate
  • Rapid rise time

    • Peak is created showing flow being delivered too fast and too high
  • Appropriate rise time
    • Just right
  • Slow rise time

    • Curve is too low
  • Modes that utilize rise time
    • PC & VC+AC
    • PC & VC+ SIMV
    • PSV
    • VS
  • Cycle asynchrony
    Occurs when the ventilator does not match the patient breathing pattern
  • Long cycling
    iT is too long, excessive and pt may attempt to actively exhale while the ventilator is still in the inspiratory phase resulting in a spike in airway pressure
  • Short (double) cycling
    iT is too short due to premature ventilator cycling
  • The parameter that allows the patient to exhale while they are on PSV mode is 25%
  • Mode asynchrony
    Occurs when the mode of ventilation selected is unable to match pts spontaneous vent pattern