MECH VENT 2
Review 1.3
Save
Share
Learn
Content
Leaderboard
Learn
Created by
Jax Sanchez
Visit profile
Cards (19)
Flow asynchrony
Inspiratory gas flow from the vent is unable to match the patients
inspiratory
flow demand
View source
Flow asynchrony
Scoop waveform
View source
VC/AC mode
Ventilator
mode where flow
asynchrony
is commonly seen
View source
Increase the
inspiratory
flow
Decrease
in I time and
increase
in E time
View source
Decrease the
inspiratory
flow
Increase
in I time and
decrease
in E time
View source
For adults, peak flow values are generally kept between
60-80
L/min
View source
For adults, the typical I-time is
0.8
to
1
second
View source
Square waveform
Lower peak flow and
decreased
inspiratory time, compared to
decelerating
View source
Decelerating waveform
Mimics more natural breathing pattern, produces high initial peak flow but
increased
iT, pts are usually placed on this waveform on
VC AC
mode
View source
Rise time
Allows
adjustment
of the
rate
View source
Rapid rise
time
Peak
is created showing
flow
being delivered too fast and too high
View source
Appropriate rise time
Just right
View source
Slow
rise time
Curve is too low
View source
Modes that utilize
rise time
PC & VC+AC
PC & VC+ SIMV
PSV
VS
View source
Cycle
asynchrony
Occurs when the
ventilator
does not match the patient
breathing
pattern
View source
Long cycling
iT is too long, excessive and pt may attempt to actively
exhale
while the
ventilator
is still in the inspiratory phase resulting in a spike in airway pressure
View source
Short (double) cycling
iT is too short due to
premature ventilator cycling
View source
The parameter that allows the patient to exhale while they are on PSV mode is
25%
View source
Mode asynchrony
Occurs when the mode of ventilation selected is
unable
to match pts
spontaneous
vent pattern
View source