The mass Spectrometer can be used to determine all the isotopes present in a sample of an element and to therefore identifyelements
The mass Spectrometer needs to be under a vacuum otherwise air particles would ionise and register on the detector
There are 4 essential steps in a mass Spectrometer
Ionisation
Acceleration
Flighttube / Iondrift area
Detection area
Ionisation
The sample can be ionised in a number if ways. Two of these techniques are electronimpact and electrosprayionisation
What are the two ionisation techniques?
Electron impact and electrospray ionisation
Electron impact
A vapourised sample is injected at low pressure
An electronic gun fires high energy electrons at the sample
This knocks out an outer electron
Forming positive ions with different charges e.g Ti -> Ti+ + e-
Electo spray ionisation
The sample is dissolved in a volatile, polar solvent
Injected through a fine hypodermic needle giving a fine mist or aerosol
The tip of needle has high voltage
At the tip of the needle the sample molecule, M gains a proton H+ from the solvent forming MH+
M(g) + H+ -> MH+(g)
The solvent evaporates away while the MH+ ions move towards a negative plate
Electron impact is used for elements and substances with low formula masselectron impact can cause larger organic molecules to fragment
Electrosorayionisation is used preferably for larger organic molecules. The softer conditions of this technique mean fragmentation does not occur
Acceleration
Positive ions are accelerated by an electric field to a constant kinetic energy
KE(J) = ½ × m(Kg) × v²(ms-1)
This equation rearranged gives
V = square root of 2KE / m
(you don't need to learn this equation but will get asked to use them in a calculation)
Acceleration
Given that all the particles have the same kinetic energy, the velocity of each particle depends on its mass. Lighter particles have a faster velocity, and heavier particle have a slower velocity
Flight tube / Ion drift area
The positive ions with smaller m/z values will have the same kinetic energy as those with larger m/z and will move faster
The heavier particles take longer to mover through the drift area
The ions are distinguished by different flight times
t= d/v. t = time if flight (s) d = length of flight tube (m) v = velocity of the particle (ms-1)
Combining the two equations gives you
t= d × squareroot of m/2KE
Detection
The ions reach the detector and generate a small current, which is fed to a computer for analysis.
The current is produced by electrons transferring from the detector to the positive ions.
The size of the current is proportional to the abundance of the species
For each isotope the mass Spectrometer can measure a m/z (mass/charge ratio) and an abundance
Sometimes two electrons may be removed from a particle forming a 2+ ion. ²⁴Mg²+ with a 2+ charge would have a m/z of 12