Keratometry and Topography

Cards (28)

  • What is the major refractive element of the eye?
    The anterior ocular surface with tear film
  • Why is measuring corneal shape important?
    • Screening tool
    • Contact lens fitting and evaluation
    • Stability of cornea and tear film
    • Aids in diagnosis/monitoring of corneal pathology
    • Gold standard for pre- and post-refractive surgery evaluation
    • Estimating astigmatism when refractive results are poor
    • IOL determination for cataract surgery
  • What is the most common instrument used to quantify corneal shape?
    Keratometer
  • What does a keratometer measure?
    Anterior corneal radius of curvature
  • How is corneal power calculated?
    By change in refractive index divided by radius of curvature
  • What is the formula for corneal power in diopters (D)?
    K (diopters) = (1.3375 - 1) / r
  • What does a K value less than 41.0 D indicate?
    A flat cornea
  • What is indicated by a K value greater than 46.0 D?
    A steep cornea
  • What are the limitations of modern keratometry?
    • Assumes refractive index = 1.3375
    • Assumes cornea is spherical or toric
    • Assumes cornea is symmetrical
    • Estimates average curvature of central 3 mm
    • Does not measure peripheral cornea
    • Limited in measuring irregular corneal surfaces
  • What was the first known keratoscope target described?
    In 1619 by Scheiner
  • What did Cuignet develop in 1820?
    The keratoscope
  • What was the purpose of the ophthalmalometer developed by Helmholtz in 1854?
    To measure the local slope of the cornea
  • What concept did Placido develop in the 1800s?
    Using a disc with concentric rings
  • What are the two systems of videokeratoscopy?
    • Placido disc system
    • Elevation based system
  • How does the Placido disc system work?
    It projects illuminated annular rings onto the cornea
  • What does the elevation based system measure?
    Anterior and posterior corneal shape and corneal thickness
  • What technology does the Pentacam use?
    Rotating Scheimpflug camera
  • What does the Orbscan combine?
    Slit scanning and Placido system
  • What does the Visante system combine?
    OCT and Placido system
  • What are the differences between axial and tangential maps in keratometry?
    • Axial: Measures curvature at a point relative to the center, simpler, less accurate
    • Tangential: Measures curvature relative to other points, more accurate, better for defects
  • What do the terms absolute and normalized refer to in keratometry maps?
    • Absolute: Same scale for every map, easy comparison
    • Normalized: Different scale for each map, adapts to range of corneal power
  • What do shape factor and eccentricity measure in corneal analysis?
    • Shape Factor: Corneal asphericity; negative SF indicates flatter center
    • Eccentricity: 0.0 = circle, 1.0 = maximum flattening; average = 0.55
  • What does the Corneal Irregularity Measurement (CIM) indicate?
    • Represents corneal surface irregularity
    • Predicts irregular astigmatism or visual distortions
    • Higher CIM values indicate higher probability of ocular pathology
  • What does Toric Keratometry Mean measure?
    • Uses elevation data to compare toric reference to actual cornea
    • Provides accurate toric representation of a patient's cornea
  • What is the effect of Excimer PRK on corneal power?
    • Initial reduction in corneal power (~4D)
    • Regression over 6 months to within 1D of original power
  • What are the ranges for normal, borderline, and abnormal values in Toric Keratometry Mean?
    • Normal: 43.10D to 45.90D
    • Borderline: 41.80D - 43.00D and 46.00D - 47.20D
    • Abnormal: 36.00D - 41.70D and 47.3D - 60.0D
  • What types of astigmatism are listed in the study material?
    1. Asymmetrical astigmatism
    2. Against the rule astigmatism
    3. Irregular astigmatism
    4. Oblique astigmatism
    5. With the rule astigmatism
  • What are the effects of scarring on corneal power after Excimer PRK?
    • Initial reduction in corneal power (~4D)
    • Regression over 6 months to within 1D of original power