urinalysis

Subdecks (2)

Cards (385)

  • quality assessment refers to overall process of guaranteeing quality patient care and is regulated throughtout the total testing system
  • QA is the continual monitoring of the entire test process from test ordering and specimen collection through reporting an interpreting results
  • pinakagoal ng quality assessment is to ensure the validity of the result
  • quality system refers to all of the laboratory's policies, processes, procedures, and resources needed to achieve quality testing
  • TAT (turnaround time) is defined as the amount of time required from the patient at which a test is ordered by the health-care provider until the results are reported to the health care-provider.
  • examination variables are the processes that directly affect the testing of specimens.
  • preexamination variables occur before the actual testing of the specimen.
  • reagent strip should be checked against known negative and positive control solutions on each shift or at a minimum once a day, and whenever a new bottle is opened.
  • instrumentation and equipment are most frequently encountered instruments; refractometers, osmometers, automated reagent strips reader, and automated microscopy instruments.
  • no patient's testing may be performed until QC is acceptable.
  • temperatures should be taken daily and recorded.
  • calibration of centrifuges performed every 3 months.
  • microscopes are kept clean at all times and have an annual proifessional cleaning.
  • deionized water used for reagent preparation is quality controlled by checking pH and purity meter resistance on a weekly basis and the becaterial count on a monthly schedule.
  • quality control refers to the materials, procedure, and techniques that monitor the accuracy, precision, and the reliabilityof a laboratory test.
  • quality control ensures that acceptable standards are met during the process of patient testing
  • QC materials should be realistic and detects critical value.
  • integrity of the specimen is not checked by the Quality Control materials.
  • external quality control are used to verify the accuracy and precisionof a test and are exposed to the same conditions as the patient samples.
  • reliability is the ability to maintain both precision and accuracy
  • mean is the average of all the data points
  • standard deviation is a measurement statistic that describes the average distance each data point in a normal point in a normal distribution is from the mean.
  • coefficient of variation is the SD expressed as a percentage of the mean.
  • chard used in microscopy: Levy jennings
  • control range is determined by setting confidence limits that are within +/-2 SD (95.5%) or +/-3SD (99.7%) of the mean.
  • trend is the gradual changing in the mean in one direction.
    shift is the abrupt change in the mean.
  • internal/procedural quality control monitors the sufficient addition of a patient specimen or reagent, the instruments/reagents interaction and for lateral flow test methods.
  • proficiency testing (PT or EQA) is the testing of unkown samples received from an outside agency, and provides unbiased validation of the quality of patient test results.
  • postexamination variables are processes that affect the reporting of results and correct interpretation of data
  • electronic transmission is now the most common method for reporting results.
  • delta check is used to discover errors and compares the patient's test results with the previous results.
  • renal functions:
    • renal blood flow
    • glomerular filtration rate
    • tubular reabsorption
    • tubular secretion
  • presence of RBC is one of the most common cause of abnormal urine color
  • color represents the concentration of the urine
  • urochrome is the product of endogenous metabolism and presence of this suggests thyroid conditions, fasting, long standing specimen
  • thundicum discovered urochrome
  • uroerythrin is the pink pigment most evident in refrigerated specimens that attaches to urate and uric acid
  • urobilin is an oxidation product of urobilinogen
  • urobilin color = orange-brown
  • urobilin indicates old specimen