  • Erlenmeyer Flask● Features slanted sides and narrow neck● Useful for mixing without spilling as well as boiling liquids
    • should not be used for precise measurements
  • Glass beaker
    • same width at top and bottom
    • useful for holding liquids
    • should not be used for precise measurements
  • Graduated cylinder
    • long and slender
    • not intended for mixing, stirring, or heating liquids
    • often used for measurements with accuracy
  • To contain pipettes - (TC) quantity/volume of liquid contained in pipette is equal to the capacity printed on the pipette
  • TD pipette - (TD) quantity/volume of liquid contained in the pipette is equal to the capacity printed on the pipette minus the drainage holdback
  • Volume of TD pipette - volume of TC pipette + drainage holdback error
  • Serological pipets● User can determined volume pipetted up to the maximum capacity of thepipet.● Common sizes are 1, 5, and 10 ml● May be TC or TD
    • Etched with graduated measurements
  • Volumetric pipettes
    • are designed to only measure 1 designated volume
    • also called "bulb" or "belly" pipettes
    • may be TC or TD
  • MLA pipette
    • commonly used in laboratory science
    • can dispense very small, but accurate volumes of liquid
    • MLA pipettes may have a designated volume or a changeable volume
  • assays
    glucose oxidase
    • more specific than copper sulfate
    • read spectrophotometrically at the endpoint of reaction, measure absorbance, generally read in visible spectrum (400-700 nm); depends on chromagen used
    • interferences - acetaminophen, excess uric acid, bilirubin, and ascorbic acid
  • QC run every 24 hours is most common
  • lyophilized - usually comes with buffer to reconstitute; dehydrated and frozen
  • types of QC mats: lyophilized, frozen, ready-to-use
  • assayed: comes form the manufacturer with a set mean and standard deviation
  • unassayed: your lab must perform test runs of this QC to establish a mean and standard deviation
  • control vials should be labeled with open date, expiration date, and initials
  • new control should be run in parallel with old control for 30 days
  • percentile: the value below which a percentage of data falls
  • standard deviation is a measure of how idspersed data is in relation to a mean
  • low SD - data is close to the mean
  • high SD - data is far from the mean
  • random error - occurs because of unpredictable events; will probably correct if repeated
  • systematic errors occur because of a problem with the system
  • examples of systematic error include shift and trend
  • shift - sudden change in values
  • shifts occur due to switching to new lot # of reagent or control
  • trend is a gradual increase or decrease in values
  • trends occur due to deterioration of the controls/reagents or contamination
  • troubleshooting - consideration of why errors occur and how to fix them
  • coefficient of variation is used to compare variation of distributions with different units relative to their means
  • generally, CV should be 5% or less
  • CV of >10% may indicate poor method performance
  • 1(2s) means that one data point is greater than 2 SD from teh mean
  • 22s is when two consecutive data points are greater than 2 SD away from the mean
  • 13S is when one data point is greater than 3SD away from the mean
  • 41s is when four consecutive data points are greater than 1 SD away from the mean
  • 10x is when ten consecutive data points fall to one side of the mean
  • R4s is when there are more than 4SD between two consecutive data points
  • ortho-toluidine is read at 630 nm
  • ortho-toluidine
    A) ortho-toluidine
    B) glucose
    C) hot acetic acid
    D) blue-green