ANACHEM II UNIT I

    Cards (46)

    • Classical/Wet-Chemical Methods
      • separating the components of interest (the
      analytes) in a sample by precipitation,
      extraction, or distillation
    • Instrumental Methods
      • make use of instruments for separating and
      determining chemical species
    • Emission of Radiation Emission Spectroscopy (X-ray, UV, Visible, Electron);
      Fluorescence, Phosphorescence, Luminescence (X-ray,
      UV, Vis, IR)
    • Absorption of Radiation Spectrophotometry & Photometry (X-ray, UV, Vis, IR);
      Photoacoustic Spectroscopy; Nuclear Magnetic
      Resonance (NMR) & Electron Spin Resonance
      Spectroscopy
    • Electrical Potential Potentiometry;
      Chronopotentiometry
    • Electrical Charge Coulometry
    • Electrical Current Amperometry; Polarography
    • Electrical Resistance Conductometry
    • Precision
      • The degree of mutual agreement among
      data that have been obtained in the same
      way
      • Provides a measure of the random or
      indeterminate error of an analysis
      • Absolute standard deviation, relative
      standard deviation, coefficient of variation,
      variance, standard deviation of the mean
    • Bias (Accuracy)
      • Provides a measure of the systematic or
      determinate error of an analytical method
      • Involves analyzing one or more standard
      reference materials whose analyte
      concentration is known
      Bias = μ – xt
    • Sensitivity
      • A measure of the instrument or method’s
      ability to discriminate between small
      differences in analyte concentration
      • Two factors that limit sensitivity:
      1. slope of the calibration curve
      2. Reproducibility or precision of the
      measuring device
    • Calibration Sensitivity
      • The slope of the calibration curve at the
      concentration of interest
    • Detection Limit
      • The minimum concentration of analyte that can be
      detected at a known confidence interval
      • Depends upon the ratio of the magnitude of the
      analytical signal to the size of the statistical
      fluctuations (standard deviation) of the blank
      • Unless the analytical signal is larger than the blank
      by some multiple of k of the variation in the blank
      owing to random errors, it is impossible to detect
      the analytical signal with certainty
    • Limit of Detection (LOD)
      • Typically 3 times the signal-to-noise
      (based on standard deviation of the noise)
    • Dynamic Range
      • Extends from the lowest concentration at
      which quantitative measurements can be
      made (limit of quantitation, LOQ) to the
      concentration at which the calibration
      curve departs from linearity (limit of
      linearity, LOL)
    • identify
      A) 1
      B) 2
      C) 3
      D) 4
    • Selectivity
      • Refers to the degree to which the method
      is free from interference by other species
      contained in the sample matrix
    • Robustness free from chemical interferences
      • can be applied to analytes in a wide variety of matrices
    • Rugged method
      • insensitive to changes in experimental conditions
    • All instrumental methods require
      calibration curve to relate measured signal
      to concentration (amount) of analyte.
    • Calibration – the process of ensuring that the signal
      measured by a piece of equipment or an
      instrument is correct
    • Standardization – the process of establishing the
      relationship between the amount of analyte and a
      method’s signal
    • Calibration Curve – the result of standardization
      showing graphically how a method’s signal changes
      with respect to the amount of the analyte
    • Primary reagent
      • Reagent of known purity that can be used
      to make a solution of known concentration
    • Secondary reagent
      • A reagent whose purity must be relative to
      a primary reagent
    • Reagent grade
      • reagents conforming to standards set by
      the American Chemical Society
    • Single-point standardization
      • A single standard containing a known
      concentration of analyte, Cs , is prepared
      and its signal, Sstand , is measured
      • The value of k is calculated as
      k = Sstand / Cs
    • Multiple-point standardization
      • Most commonly employed standardization
      method
      • Uses one or more external standards
      containing known concentration of analyte
    • External standards - prepared and
      analyzed separately from the samples
    • Three Types of Calibration Curves
      1. External Standard
      2. Standard Addition
      3. Internal Standard
    • External Standard
      • Constructed by measuring signals
      over several known concentrations of
      analyte and then plotting signal vs
      concentration
      • Known analyte solutions must have
      similar matrix to that of sample
    • External Standard
      Signal magnitude should not be a function
      of small sample volumes that may be
      difficult to reproduce
      • Works well for simple matrices
      • Calibration curve prepared from known
      concentration
      • Unknown solution measured
    • Standard Addition Known amounts of analyte are added
      directly to sample to try to account for
      matrix effects
      • Two main types:
      Constant Total Volume
      Changing Total Volume'
    • Standard Addition: Constant Total Volume
      Analysis conditions:
      – Constant volume of unknown analyte
      – Varying volume of added analyte of known
      concentration to each sample
      – Linear response of instrument to
      concentration
    • Standard Addition: Constant Total Volume
    • Two-point Standard Addition:
      Constant Total Volume
      Can be done with only two points using
      the unknown and one addition to sample
    • Two-point Standard Addition:
      Changing Total Volume
      Signal from known volume of unknown
      concentration is measured (S1
      • Signal after known volume of standard
      added to the unknown sample is measured
      (S2
    • Two-point Standard Addition:
      Changing Total Volume
      Signal from known volume of unknown
      concentration is measured (S1)
      • Signal after known volume of standard
      added to the unknown sample is measured
      (S2
      )
    • How to Determine if Standard Addition is Needed
      • Have to do both ways and see if external
      standard gives numbers that are always too
      high or too low.
      • External standard
      faster
      – More efficient if many samples and fast
      analysis time for each sample is needed
      • Standard Addition
      • Used only if external standards do not give
      correct answers
    • Voltammetric Analyses almost always
      use standard addition because:
      each measurement that requires
      emptying and refilling cell takes
      several minutes
      addition of known to solution that
      has already been deoxygenated
      allows for quick measurement
      • electrode surface is very sensitive
      to changes in solutions