ANACHEM II UNIT I

Cards (46)

  • Classical/Wet-Chemical Methods
    • separating the components of interest (the
    analytes) in a sample by precipitation,
    extraction, or distillation
  • Instrumental Methods
    • make use of instruments for separating and
    determining chemical species
  • Emission of Radiation Emission Spectroscopy (X-ray, UV, Visible, Electron);
    Fluorescence, Phosphorescence, Luminescence (X-ray,
    UV, Vis, IR)
  • Absorption of Radiation Spectrophotometry & Photometry (X-ray, UV, Vis, IR);
    Photoacoustic Spectroscopy; Nuclear Magnetic
    Resonance (NMR) & Electron Spin Resonance
    Spectroscopy
  • Electrical Potential Potentiometry;
    Chronopotentiometry
  • Electrical Charge Coulometry
  • Electrical Current Amperometry; Polarography
  • Electrical Resistance Conductometry
  • Precision
    • The degree of mutual agreement among
    data that have been obtained in the same
    way
    • Provides a measure of the random or
    indeterminate error of an analysis
    • Absolute standard deviation, relative
    standard deviation, coefficient of variation,
    variance, standard deviation of the mean
  • Bias (Accuracy)
    • Provides a measure of the systematic or
    determinate error of an analytical method
    • Involves analyzing one or more standard
    reference materials whose analyte
    concentration is known
    Bias = μ – xt
  • Sensitivity
    • A measure of the instrument or method’s
    ability to discriminate between small
    differences in analyte concentration
    • Two factors that limit sensitivity:
    1. slope of the calibration curve
    2. Reproducibility or precision of the
    measuring device
  • Calibration Sensitivity
    • The slope of the calibration curve at the
    concentration of interest
  • Detection Limit
    • The minimum concentration of analyte that can be
    detected at a known confidence interval
    • Depends upon the ratio of the magnitude of the
    analytical signal to the size of the statistical
    fluctuations (standard deviation) of the blank
    • Unless the analytical signal is larger than the blank
    by some multiple of k of the variation in the blank
    owing to random errors, it is impossible to detect
    the analytical signal with certainty
  • Limit of Detection (LOD)
    • Typically 3 times the signal-to-noise
    (based on standard deviation of the noise)
  • Dynamic Range
    • Extends from the lowest concentration at
    which quantitative measurements can be
    made (limit of quantitation, LOQ) to the
    concentration at which the calibration
    curve departs from linearity (limit of
    linearity, LOL)
  • identify
    A) 1
    B) 2
    C) 3
    D) 4
  • Selectivity
    • Refers to the degree to which the method
    is free from interference by other species
    contained in the sample matrix
  • Robustness free from chemical interferences
    • can be applied to analytes in a wide variety of matrices
  • Rugged method
    • insensitive to changes in experimental conditions
  • All instrumental methods require
    calibration curve to relate measured signal
    to concentration (amount) of analyte.
  • Calibration – the process of ensuring that the signal
    measured by a piece of equipment or an
    instrument is correct
  • Standardization – the process of establishing the
    relationship between the amount of analyte and a
    method’s signal
  • Calibration Curve – the result of standardization
    showing graphically how a method’s signal changes
    with respect to the amount of the analyte
  • Primary reagent
    • Reagent of known purity that can be used
    to make a solution of known concentration
  • Secondary reagent
    • A reagent whose purity must be relative to
    a primary reagent
  • Reagent grade
    • reagents conforming to standards set by
    the American Chemical Society
  • Single-point standardization
    • A single standard containing a known
    concentration of analyte, Cs , is prepared
    and its signal, Sstand , is measured
    • The value of k is calculated as
    k = Sstand / Cs
  • Multiple-point standardization
    • Most commonly employed standardization
    method
    • Uses one or more external standards
    containing known concentration of analyte
  • External standards - prepared and
    analyzed separately from the samples
  • Three Types of Calibration Curves
    1. External Standard
    2. Standard Addition
    3. Internal Standard
  • External Standard
    • Constructed by measuring signals
    over several known concentrations of
    analyte and then plotting signal vs
    concentration
    • Known analyte solutions must have
    similar matrix to that of sample
  • External Standard
    Signal magnitude should not be a function
    of small sample volumes that may be
    difficult to reproduce
    • Works well for simple matrices
    • Calibration curve prepared from known
    concentration
    • Unknown solution measured
  • Standard Addition Known amounts of analyte are added
    directly to sample to try to account for
    matrix effects
    • Two main types:
    Constant Total Volume
    Changing Total Volume'
  • Standard Addition: Constant Total Volume
    Analysis conditions:
    – Constant volume of unknown analyte
    – Varying volume of added analyte of known
    concentration to each sample
    – Linear response of instrument to
    concentration
  • Standard Addition: Constant Total Volume
  • Two-point Standard Addition:
    Constant Total Volume
    Can be done with only two points using
    the unknown and one addition to sample
  • Two-point Standard Addition:
    Changing Total Volume
    Signal from known volume of unknown
    concentration is measured (S1
    • Signal after known volume of standard
    added to the unknown sample is measured
    (S2
  • Two-point Standard Addition:
    Changing Total Volume
    Signal from known volume of unknown
    concentration is measured (S1)
    • Signal after known volume of standard
    added to the unknown sample is measured
    (S2
    )
  • How to Determine if Standard Addition is Needed
    • Have to do both ways and see if external
    standard gives numbers that are always too
    high or too low.
    • External standard
    faster
    – More efficient if many samples and fast
    analysis time for each sample is needed
    • Standard Addition
    • Used only if external standards do not give
    correct answers
  • Voltammetric Analyses almost always
    use standard addition because:
    each measurement that requires
    emptying and refilling cell takes
    several minutes
    addition of known to solution that
    has already been deoxygenated
    allows for quick measurement
    • electrode surface is very sensitive
    to changes in solutions