1. Centrifuge specimen as soon as possible after collection to obtain platelet poor plasma
2. Incubate a sufficient amount of 0.025-m calcium chloride at 37 C
3. Pipet .02 mL of control plasma (or patient’s plasma) into a test tube
4. Pipet 0.2 mL of partial thromboplastin (containing activator) into the test tube containing the control (or patient’s) plasma
5. Mix the contents of the tube quickly and place in a 37°C water bath for 5 minutes
6. After exactly 5 minutes, forcibly pipet 0.2 mL of prewarmed calcium chloride into the tube, and simultaneously start stopwatch
7. Mix the test tube immediately after adding calcium chloride. Allow the test tube to remain in the water bath while gently tilting the tube every 5 seconds. At the end of 20 seconds, remove the test tube from the water bath. Quickly wipe off the outside of the test tube with clean gauze so that the contents of the tube can be clearly seen
8. Gently tilt the test tube back and forth until clot forms, at which point the timing is stopped
9. Report the patient results in seconds along with the normal for the test as determined for your laboratory