Signal originates from the nuclei of atoms resonating in a patient in the presence of a magnetic field
Quantum mechanics
Branch of physics that describes the behavior of very small objects, such as x-rays, protons, neutrons, and electrons
Quantum mechanics deals with subatomic particles and photons
Spin
Quantity that every nucleus has, quantized into units of half-integer values and called the spin quantum number
Allowed values of spin quantum number
0
1/2
1
3/2
and so on
All hydrogen atoms of atomic mass 1 have a spin quantum number of 1/2
Spin state
Limited number of ways a nucleus can spin, either counter clockwise (+1/2) or clockwise (-1/2)
A spinning, charged mass induces a magnetic field about itself, like the Earth
Nuclear magnetic moment
The magnetic field associated with the spinning, charged nucleus, related to its mass, charge, and rate of spin
Neutrons with no charge have no associated magnetic field
Proton
Classifies what kind of atom something is, the number of protons is the atomic number
Isotope
Atoms with the same number of protons but different numbers of neutrons
Hydrogen is the most abundant atom in the human body
Equilibrium
Before being energized or receiving RF, the nuclear magnetization is aligned longitudinally with the main magnetic field
At equilibrium, the parallel aligned protons are in a low energy state
Energy cannot be created or destroyed, it can only be transformed
Antiparallel/Spin down
High energy state of the protons
In the presence of the main magnetic field, there are protons that are out of phase
Net magnetization
The overall magnetization of the protons, which is related to the external magnetic field strength
The larger the net magnetization (M0), the stronger the MR signal
Increasing the concentration of the nuclear species or the volume of tissue increases the MR signal due to higher proton density
Hydrogen
Has the highest gyromagnetic ratio of any nucleus other than tritium, and the second largest signal per number of nuclei
Tritium would make an excellent MRI agent but is radioactive
For 13C, the gyromagnetic ratio is about one fourth the value for 1H, resulting in a much weaker signal
Higher magnetic field strength (B0) results in stronger net magnetization and a stronger MR signal
Larmor frequency
The precessional frequency of the nuclei, corresponding to electromagnetic radiation in the RF range
The RF pulse must be tuned to the Larmor frequency to rotate the net magnetization vector
Transverse magnetization (Mxy)
The net magnetization in the transverse (XY) plane, which is the only magnetization that can be detected as an MR signal
It is not possible to directly measure the longitudinal magnetization (Mz)
Stationary frame of reference
The view of someone standing next to the magnet, where all motions are compared to someone standing still
Rotating frame of reference
A frame of reference that matches the motion of the net magnetization, making it easier to visualize
RF pulse
The application of a designed time/intensity radiofrequency pulse that can rotate the net magnetization vector to any angle
Hard pulse
A strong, very short RF pulse
Soft pulse
A weaker but longer RF pulse
Flip angle
The angle through which the RF pulse rotates the net magnetization vector
90 degree RF pulse
Rotates the net magnetization vector from equilibrium onto the transverse (XY) plane
180 degree RF pulse
Rotates the net magnetization vector from equilibrium to the negative Z-axis
Alpha (α) pulse
A "partial flip" RF pulse with a rotation angle less than 90 degrees
After a 90 degree RF pulse, the longitudinal magnetization (Mz) is zero and the transverse magnetization (Mxy) is equal to the equilibrium magnetization (M0)
Dephasing of Mxy
The gradual loss of phase coherence in the transverse magnetization due to interactions with local magnetic fields