Platelet aggregometry using platelet-rich plasma
1. The operator pipettes the PRP to instrument compatible cuvettes
2. Drop in one clean plasticized stir bar per sample
3. Place the cuvettes in incubation wells
4. Allows the samples to warm to 37C for 5 mins
5. The operator than transfers the first cuvette, containing specimen and stir bar, to the instrument's reaction well
6. Start the stirring device and the recording computer
7. The instrument directs focused light through the sample cuvette to a photodetector
8. The operator pipettes an agonist (platelet activator) directly into the same to trigger aggregation