A simple but important aid to viewing radiographs, commonly used throughout X-ray departments, consisting of an even light source of sufficient light intensity to allow optimum viewing
Viewing box
Conveniently placed
Safely fixed to a wall or stable mobile structure
In good working order
Of satisfactory design
Has an all over even light
Clean inside and out
Electrically safe
Has an adequate light output
Viewing box design
1. Two parallel fluorescent tubes of equal light output, with starters
2. A built in, or attached, spotlight is an advantage
3. The film anchor should hold the film firmly
4. A good quality switch
5. Safe electrical wiring
Viewingboxcleaning
1. Remove plug from power socket
2. Remove the front window
3. Clean window on both sides (intensifying screen cleaner is a useful cleaning agent as it has anti-static properties)
4. Clean the back plate and fluorescent tubes
5. Inspect window for damage
6. Make sure that tubes and starters are firmly in place
7. Replace window
8. Test
Dust on the window or tube will reduce the light output
Marks on the window can be misleading
Frequency of viewing box cleaning
Outside daily
Inside 6monthly
Equipment required for viewing box cleaning
Viewing box to be cleaned
Clean cloth
Screw driver
Electrical check should be carried out by or under the supervision of an electrician
Items to be checked during electrical check
Switches
Fluorescent tubes
Starters
Electrical connections
Frequency of electrical check
6 monthly
Equipment required for electrical check
Viewing box to be checked
Screw driver
Electricalcheck procedure
1. Remove plug from socket
2. Remove front window
3. Check condition of wiring
4. Check electrical connections
5. Check stability of switch
6. Check installation of fluorescent tubes and starters
7. Replace front window
8. Replace plug in socket
9. Test
Replacing fluorescent tubes at different time intervals, or with a different type of tube, may produce a difference in light intensity from one side of the screen to the other or one viewing box to another
If, after replacing only one tube, it is found to be brighter than the remaining one, replace both
The working tube that has been removed can be matched with another one at a later date
Room light levels in the film viewing area should not be too high
A bright spotlight should be available for viewing dark areas of the film
The viewing box should be mounted at the correct height for easy viewing
Light source should be even and the same as other boxes
Cassette Inspection Frequency
Yearly
Cassette Inspection
Inspection for faulty parts, distortion, and inadequate cleanliness of cassette
Light Leakage Test Frequency
Yearly
Light Leakage Test
Test for black fogging on edge of film
Cassette Inspection Equipment
Casettes
Light Leakage Test Equipment
Cassettes
Intensifying Screen Inspection Frequency
Monthly
Intensifying Screen Inspection Equipment
Screens
I.S. Inspection
Inspect for abrasions, foregin material on screen, discoloration
I.S. Cleaning Frequency
Monthly
I.S. Cleaning
All dust particles removed, check if there are smears