In a PGC, with a fixed sample loop and isothermal oven, the Ideal Gas Equation becomes P α n, hence the number of moles that are trapped is solely a function of the pressure in the loop
Potential problems with Plunger Diaphragm Valves are diaphragm leakages of minute quantities of liquid samples in the diaphragm material or at the points where the diaphragm is pinched
1. At sample injection, V1 is switched to allow the Carrier 1 to push the sample gas in the sample loop into Col. #1
2. After sample injection, V1 switches back to the original position and the carrier gas continues to push the sample gas into the column
3. After the sample passes through Col. #1 and the separated components flow to Col. #3, before all the separated components can pass through V2, both V2 valves switch to push the backend from Col. #1 back to Col. #1 and the frontend into Col. #3
4. After the frontend passed through Col. #3, both V2valves switch back to original position
To reduce the total analysis time by flushing unwanted heavier components to vent, to remove components that would interfere with the required separation, to remove components that would damage one of the columns, to regroup a number of components for measurement, and to maintain good performance by ensuring that the columns do not accumulate impurities from the sample
To remove most of a major component, to reduce the amount of a major component reaching the second column down to a concentration level that does not tail, to increase the amount of the measured component relative to the major component to 1 in 20 or better, and to remove other components that may interfere with the analysis
The lighter components that elute from Column 1 are allowed to flow into Column 2 and trapped, while the heavier components pass via Column 2 Simulator