Column Switching - C2 Application
1. At sample injection, V1 is switched to allow the Carrier 1 to push the sample gas in the sample loop into Col. #1
2. After sample injection, V1 switches back to the original position and the carrier gas continues to push the sample gas into the column
3. After the sample passes through Col. #1 and the separated components flow to Col. #3, before all the separated components can pass through V2, both V2 valves switch to push the backend from Col. #1 back to Col. #1 and the frontend into Col. #3