1. Volatile liquid or gaseous sample is injected through a septum into a heated port which rapidly evaporates
2. Vapor is carried through the column by the carrier gas. Separation is based on the affinity of the solute particles to the carrier gas and the stationary phase
3. Separated analytes flow through a detector, which is maintained at a temperature higher than the column. This will ensure that your analyte is still in a gaseous phase
4. The detected response is displayed in the computer as a chromatogram
1. Varying kinds of solutes react differently with stationary phase
2. Programming temperature may improve the situation
3. When oven condition is set at a particular temperature, say 150∘C, volatile compounds will elute first but the less volatile ones remain in the column
4. When temperature is increased from 50 ∘C to 250 ∘C at constant flow rate, elution of all compounds may occur uniformly
5. Retention time of the analytes will also decrease
6. The highest temperature should not be set at a point where stationary phase and some analytes would start to decompose
7. Pressure programming is also a plausible option when temperature programming is not possible, as when decomposition of analytes take place at high temperature
Very fast separation, fastest optimal flow rate with low penalty on resolution
Its drawbacks include: possible reaction with unsaturated compounds on metal surfaces, incompatibility with mass spectrometric detector, and formation of explosive mixture with air
Necessary in the analysis by GC of compounds that exhibit low volatility, poor thermal stability, contain "active" groups that can lead to loss of sample to intermolecular hydrogen bonding or adsorption on the inlet or column, and/ or demonstrate poor sensitivity at the detector