RADT Unit 4
Imaging System Characteristics
Save
Share
Learn
Content
Leaderboard
Learn
Created by
Chris Z
Visit profile
Cards (20)
Dual focus Tubes mm small and large:
small
0.5mm
0.6mm
1.0
mm
large
1.0mm
1.2mm
2.0mm
0.3/1.0mm is typically used for :
angiography
and
magnification
procedures
0.1/0.3mm is typically used for:
mammography
What is Microfocus Tubes
FSS
of
0.3mm
or
less
Large focus
normal
imaging
higher
mA
can be impressed for
larger
/
dense
body parts
What is the main advantage of large FSS in relation to detail?
Shorter exposure times
to
minimize motion
Small focus
given amount of
electrons
cover
smaller
area on
anode
cannot use
high mA
due to
limited heat capacity
used for
magnificatoin
radiography where detail is required
for
smaller
and
thinner
body parts
X-ray beam is
polyenergetic
and
heterogeneous
Filtration
hardens
the beam
makes beam more
uniform
by removing
softer
rays
What are the two classifications of filtration?
Inherent
Added
What is inherent filtration made of?
glass
or
metal
envelope (
0.5mm
Aluminum equivalent)
What is added filtration?
mirror
in
collimator
box (
1.0mm
Aluminum equivalent)
another 1.0mm AI eq added between
x-ray
tube
housing
and
collimator
What is the total of inherent and added filtration?
.5mm
of Inherent Filtration+
2.0mm
of Added Filtration=
2.5mm
of
Total
Filtration AI equivalent
Half-Value
Layer
(
HVL
)
improves
quality
of the beam
is the amount of material that reduce the
intensity
of the
x-ray
beam to
half
its original value
used for
shielding
and
filtration
increase in HVL
increases
the quality of the beam
Compensation Filter
Different shapes
of
Aluminum
mounted under the
variable light
aperture
(
collimator
)
Compensation filter
balances
intensity
of x-ray beam for a more uniformed exposure to the
image receptor
What are the two different shapes of compensation filter?
Wedge
filter
Trough
Filter
Wedge filter is good for:
body parts thick
on one end and
thin
on the other ex) foot, hips
Trough Filter is good for:
chest
Anode Heel Effect
phenomenon which
intensity
of x-ray beam is
greater
towards the
cathode side
of the tube
the loss of intensity is a result of the electrons traveling
farther
to the
anode heel
portion of the target
Where would you place the thicker part of the body?
At the
cathode
end