1. Click "QC" button on the top of the screen to enter the quality control interface
2. Adding QC File - Click the "Set-Up" then "Add" button and input the Lot number, the expiry date, target values, and limits for parameters of the control, then select "Blood Mode"
3. Editing QC File - Click the "Set-Up" button, then select a QC file and click the "Edit" button to input the Lot number, expiry date, target values, and limits for parameters of the control, then select "Blood Mode"
4. Testing QC - In the QC interface, select the corresponding QC file and test the control in the same way as testing normal samples
5. Reviewing QC Results - In "QC Table" menu, you can print, delete, or clear all QC results
6. Checking QC Graph - Click "QC Graph" button to enter the QC graph interface, the QC results are shown in the graph, determine whether they are out of control according to the L-J curve