Haemocytometer concentrations
1. 2 counting chambers w/microscopic grid
2. 3% saline is used prior to loading sample to immobilise sperm and preserve them for later counting
3. Must be dilute enough to count individual sperm (by factor of 400 suitable for ram)
4. Mix sample thoroughly before counting to evenly distribute sperm
5. Fixe coverslip over counting chambers by first moistening either side to prevent movement
6. Load into haemocytometer and place into humid chamber to let sample settle for 3-5 mins
7. 25 larger squares with 16 smaller squares in each
8. Only need to count sperm in 5 larger squares (count gained from 4 corner squares and centre square)
9. Swing in 20x or 40x objective to count sperm
10. All sperm including loose heads within the smaller squares are counted
11. Sperm falling in centre of triple ruled line are only counted if located on upper or left hand side of square
12. Sperm on lower and right hand side of square in those lines are discounted
13. Use cell counter, add counts together before progressing to 2nd counting chamber
14. If less than 10 sperm are counted in first square, all 25 squared need to be counted to reduce variability