tissue weighting factor of red bone marrow, colon, lung, and stomach:
0.12
tissue weighting factor of gonads:
0.20
the smaller the number of the tissue weighting factor, the __________ radiosensitive the tissue is
less
annual effective dose limit for occupational exposure:
50 mSv (5 rem)
cumulative effective dose limit for occupational exposure:
10 mSv (1 rem) x age in years
equivalent dose annual limit for lens of eye for occupational exposures:
150 mSv (15,000 rem)
equivalent dose annual limit for localized areas of the skin, hands, and feet for occupational exposures:
500 mSv (50,000 rem)
effective dose limit, continuous/frequent exposure, for the general public:
1 mSv (100 mrem)
effective dose limit, infrequent exposure, for the general public:
5 mSv (500 mrem)
equivalent dose limit of the lens of eye for the general public:
15 mSv
equivalent dose limit of the localized areas of the skin, hands, and feet for the general public:
50 mSv
effective dose of remedial action for natural sources, excluding radon, for the general public:
>5 mSv (>500 mrem)
exposure to radon and its decay products for the general public:
>26 (J/s)m^3
effective dose limit for education and training exposures annually:
1 mSv (100 mrem)
equivalent dose limit for annual education and training exposures to lens of eye:
15 mSv
equivalent dose limit for annual education and training exposures to localized areas of skin, hands, and feet:
50 mSv
monthly equivalent dose limit for embryo-fetus exposures:
0.5 mSv
equivalent dose limit for embryo-fetus exposures during the entire gestation period:
5.0 mSv
negligible individual dose annual limit:
0.01 mSv (1 mrem)
early and late tissue reactions have a high probability of occurring when entrance radiation doses exceed ________
2 Gyt
SID indicator must be accurate to within _________ of the actual SID
2%
the collimation beam and xray beam must coincide to within ________ of the SID
2%
the PBL must be accurate to within _________ of the SID
2%
all general purpose diagnostic xray beams must have a total filtration of at least _________ when operated above __________
2.5 mm Al; above 70 kVp
radiographic tubes operated between 50-70 kVp must have at least __________ of filtration
1.5 mm Al
below 50 kVp, a minimum of _________ of total filtration is required
0.5 mm Al
kilovoltage accuracy must be accurate within __________ of the control panel setting, which equates to about + or - _________ kV of the kV selected
10%; + or - 5 kV
variation in xray intensity should not exceed ________
5%
the maximum acceptable variation in linearity is ________ from one mA station to an adjacent mA station
10%
if the backup timer is controlled automatically, it should terminate at a maximum of ___________ when equipment is operated at or above ________
600 mAs; 50 kVp
backup time should be set at ________-________ of the expected exposure time
150-200% (or 1.5 to 2 times)
the annual nontherapeutic medical radiation dose to the US population in 2006 was _________ and decreased to _________ in 2016
2.92mSv decreased to 2.16mSv
lifetime effective dose of radiation workers must be limited to:
individual's age in years times 10 mSv
epidemiologic studies of atomic bomb survivors exposed in utero provided conclusive evidence of a dose dependent increase in the incidence of severe intellectual disability for fetal doses higher than approximately ________
0.4 Sv
the limit for any education and training exposures of individuals under the age of 18 years old is an EfD of ______ annually
1 mSv
the CumEfD for a 90 year old is:
900 mSv
2 things that every diagnostic imaging system must have:
protective tube housing
correctly functioning control panel
distance indicators must be accurate to within ________ of the SID