Mod 4

    Cards (61)

    • 3 commonly used radiotherapy technologies
      • External beam radiation therapy (EBRT or XRT or teletherapy)
      • Brachytherapy or sealed source radiation therapy
      • Systemic radioisotope therapy or unsealed source radiotherapy
    • Radioactivity
      A process whereby a nucleus that contains an excess energy undergoes a transformation to a more stable state by emitting energy in the form of particles or electromagnetic radiation
    • Radioactive half-life
      The time interval in which the original activity level, tabulated as disintegrations per second (dps) decreases by one half
    • Radionuclide/Radioisotopes and their half-lives
      • Polonium-215 (0.0018 sec)
      • Bismuth-212 (60.5 sec)
      • Sodium-24 (15 hrs)
      • Iodine-131 (8.07 days)
      • Cobalt-60 (5.26 yrs)
      • Radium-226 (1,600 yrs)
      • Uranium-238 (4.5 billion yrs)
    • Superficial X-Ray tube (Philips RT 100)
      • X-ray produced at 50-150 kV
      • Varying thickness of filtration (usually 1-6 mm Al) are added to harden the beam
      • Superficial treatments are usually given with the help of applicators or cones
      • SSD range 15 to 20 cm
      • Dose is highly dependent on source-skin distance, filtration and applicator area
      • Usually operated at 5-8 mA
      • Useful for irradiating tumor confined to about 5 mm depth
    • Issues with superficial radiotherapy
      • Short focus to skin distance (FSD) and hence high output and large influence of inverse square law
      • Calibration difficult due to strong dose gradient i.e. dose fall off and electron contamination
    • Low-energy machines
      • Superficial Equipment
      • Orthovoltage Units
    • Grenz rays
      • 10-15 kVp
      • Treatment of inflammatory disorders (Langerhans' cells), Bowen's disease, patchy stage mycosis fungoides, herpes simplex
    • Contact therapy
      • Superficial skin lesions
      • Endocavitary treatments for curative intent (rectal)
    • Superficial equipment
      • 50-150 kVp
      • Skin cancer and tumors no deeper than 0.5 cm
    • Orthovoltage machines
      • 150-500 kVp
      • Skin, mouth, and cervical carcinoma
      • Experience limitation in the treatment of lesions deeper than 2 to 3 cm
    • Orthovoltage units
      • Uses "conventional" X-Ray tube with electrons accelerated by an electric field
      • Stationary anode (in contrast to diagnostic tubes which have a rotating anode)
      • Filtration important
    • Limitations of low energy machines
      • Can not reach deep-seated tumors with an adequate dosage of radiation
      • Do not spare skin and normal tissues
    • Differences between superficial and orthovoltage
      • Superficial: 50 to 150kVp, small skin lesions, maximum applicator size typically < 7cm, typical FSD < 30cm, beam quality measured in HVL aluminium (0.5 to 8mm)
      • Orthovoltage: 150 to 500kVp, skin lesions, bone metastases, applicators or diaphragm, FSD 30 to 60cm, beam quality in HVL copper (0.2 to 5mm)
    • Orthovoltage units

      • Uses conventional X-ray tube
      • Energy range 150- 500 kV X-rays
      • Mostly used around 250 - 300 kVp
      • Applicators are used in superficial therapy
      • Treatment depths of around 20 mm
      • Penetration sufficient for palliative treatment of bone lesions relatively close to the surface (ribs, spinal cord)
    • Disadvantages of deep X-ray
      • Higher dose to bone - photoelectric absorption
      • Maximum dose on the surface hence higher skin dose
      • Treatment to a depth of only a few centimeters possible
      • High scattered radiation and larger penumbra
    • Orthovoltage X-ray filters
      • Copper and sometimes Tin in addition to Aluminum
      • Aluminum is placed distal to copper to remove soft secondary radiation
      • Copper is placed distal to Tin to remove soft secondary radiation when beam interacts with Tin
    • Orthovoltage X-ray operation
      • Usually operates at SSD (Source to surface distance) 50 - 70 cm with or without cone
      • Has moveable lead shields
      • Low dose rate due to long SSD and heavy filtration of beam
    • Superficial machine or kilovoltage units
      • Operating with X-rays tubes at an accelerating potential < 50 kV
      • Superficial therapy refers to potentials of 50-150
      • Low/poor penetrating ability
      • Additional filters in the form of aluminum, tin and copper are added for beam hardening
      • Used in treatment of superficial lesions
      • Cone is directed to skin surface of patient
      • Lead shielding to be placed directly to shield surrounding areas
      • Treatment distance is usually 15 - 20 cm to decrease depth dose
      • Higher patient dose due to distance
      • Backscatter is high due to low energy and increases with increased field distance
    • Orthovoltage therapy or deep therapy
      • Energy: 200 - 300 kV
      • Tube current: 10 - 20 mA
      • HVLs: 1 - 4 mm Cu
      • Cones or movable diaphragm (continuous adjustable field size)
      • SSD: 50 cm
      • Application: tumor located < 2-3 cm in depth
    • Limitations of orthovoltage therapy
      • Skin dose
      • Depth dose distribution
      • Increase absorbed dose in bone
      • Increase scattering
    • Contact X-ray brachytherapy
      • Also called "CXB", "electronic brachytherapy" or the "Papillon Technique"
      • A type of radiation therapy using X-rays applied close to the tumour to treat rectal cancer
    • Megavoltage
      A clinical modality consisting of the administration of high energy (1 megavolt or greater) to a deep-seated cancer (e.g., prostate or brain cancer) or cerebrovascular defects by an MRT unit (e.g., linear accelerator or 60Co unit)
    • Radioisotope therapy
      • Delivers radiation directly into the cancer cells, usually as a capsule, drink or injection into a vein
      • Cancer cells absorb the radioactive substance more than normal cells, receiving a higher dose of radiation and causing the cells to die
    • Betatron
      • A cyclic accelerator that produces high-energy electrons for radiotherapy
      • The magnetic field of the betatron deflects electrons into a circular orbit, and an increasing magnetic orbital flux produces an induced circumferential electric field that accelerates them
      • Developed by KERST in 1941
      • Replaced by LINAC in 1950
    • Disadvantages of betatron
      • Low dose
      • Limited field size
      • Needs large treatment room due to large size
      • Limited motion
    • Van de Graaff generator
      • An electrostatic generator which uses a moving belt to accumulate electric charge on a hollow metal globe on the top of an insulated column, creating very high electric potentials
      • Produces very high voltage direct current (DC) electricity at low current levels
      • 1931 (MIT)
      • 40 feet high Electrostatic device capable of operating at 5,000,000 volts
      • 2 MeV Clinical Van de Graaff X-ray machine
      • Invented by Robert van de Graaff
    • Machines using isotopes
      • Teletherapy - Treatment in which source of radiation is at some distance from the patient
      • Brachytherapy - The source of radiation is very close to treated tissue
    • Cobalt-60
      • On and off shield source
      • Emits two photons/disintegration: 1.17 and 1.33 Mev which is useful in radiation therapy
      • The dose rate is constantly decreasing and adjustment of treatment is done periodically
      • Large size: penumbra
      • LINAC: small focal spot
      • Cobalt: near: small source: activity and dose is reduced
    • Cesium-137
      • Teletherapy isotope machine
      • Same with the shield source of Cobalt-60
      • 20 to 30 cm with 0.662 MeV
    • Radioactive sources (γ ray equipment)
      • Cobalt 60
      • Cesium 137
    • Telecurie units
      • Cs-131
      • Cobalt-60 Unit
    • Features of a teletherapy source
      • High energy gamma ray emission
      • High specific air kerma rate constant
      • Simple means of production
      • Low cost
      • High specific activity
      • Long half-life
    • Cesium-137
      • Photon energy 0.66MeV
      • Relatively large source to relatively low specific activity
      • Medium FSD (around 60cm)
      • No isocentric mounting - similar to orthovoltage equipment in set-up
      • Not sold anymore and should not be in use
    • Natural cobalt (59Co)

      • Kobald, from the German for goblin or evil spirit
      • Discovered in 1735
      • Brittle hard metal similar to iron and nickel
      • Found in minerals and meteorites
      • Salts and glass oxides are deep blue in colour
    • Differences between orthovoltage and telecobalt unit
      • Orthovoltage: 150-500 KV x-rays, maximum dose on the skin, treatment to a depth of few centimeters, higher absorption by bone, non uniform dose distribution, higher side scatter hence larger penumbra, mostly isocentric unit
      • Telecobalt: 1.25 MeV γ Photon, maximum dose at depth of 5 mm, relatively uniform dose absorption, higher penetration deep seated tumours, more of forward scatter, lesser penumbra
    • Differences between linear accelerator and telecobalt unit
      • Linear Accelerator: 4 to 21 MV photon beams, maximum dose at higher depth with energy, no radioactive source, radiation only when the source is switched is ON, uniform dose absorption, 1mm source - nearly point source, small Penumbra, electron beam of various energies possible
      • Telecobalt: 1.25 MeV γ Photon, maximum dose at depth of 5 mm, source to be changed every 4 to 5 years, leakage radiation present even while the beam is off, relatively uniform distribution, 1-2 cm source diameter, larger penumbra, gamma Photon only
    • Telecobalt unit
      • Introduced in the 1950's, being replaced by linacs
      • The first practical radiation therapy treatment unit to provide a significant dose below the skin surface and simultaneously spare the skin
      • Still used in developing countries: simpler design, cost, little tech support
    • Properties of telecobalt unit
      • Photon energy around 1.25MeV
      • Specific activity large enough for FSD of 80cm or even 100cm
      • Therefore, isocentric set-up possible
      • Constantly emit radiation
      • 60Co source must be shielded in a protective housing (source head)
      • Source head is a steel shell filled with lead (may be up to 2 ft in diameter)
    • Application of telecobalt unit
      • To treat cancers of the head and neck area, breast, spine, and extremities
      • Areas just below the skin surface