A process whereby a nucleus that contains an excess energy undergoes a transformation to a more stable state by emitting energy in the form of particles or electromagnetic radiation
Superficial: 50 to 150kVp,small skin lesions, maximum applicator size typically < 7cm, typical FSD < 30cm, beam quality measured in HVL aluminium (0.5 to 8mm)
Orthovoltage: 150 to 500kVp, skin lesions, bone metastases, applicators or diaphragm, FSD 30 to 60cm, beam quality in HVL copper (0.2 to 5mm)
A clinical modality consisting of the administration of high energy (1 megavolt or greater) to a deep-seated cancer (e.g., prostate or brain cancer) or cerebrovascular defects by an MRT unit (e.g., linear accelerator or 60Co unit)
A cyclic accelerator that produces high-energy electrons for radiotherapy
The magnetic field of the betatron deflects electrons into a circular orbit, and an increasing magnetic orbital flux produces an induced circumferential electric field that accelerates them
An electrostatic generator which uses a moving belt to accumulate electric charge on a hollow metal globe on the top of an insulated column, creating very high electric potentials
Produces very high voltage direct current (DC) electricity at low current levels
1931 (MIT)
40 feet high Electrostatic device capable of operating at 5,000,000 volts
Differences between orthovoltage and telecobalt unit
Orthovoltage: 150-500 KV x-rays, maximum dose on the skin, treatment to a depth of few centimeters, higher absorption by bone, non uniform dose distribution, higher side scatter hence larger penumbra, mostly isocentric unit
Telecobalt: 1.25 MeV γ Photon, maximum dose at depth of 5 mm, relatively uniform dose absorption, higher penetration deep seated tumours, more of forward scatter, lesser penumbra
Differences between linear accelerator and telecobalt unit
Linear Accelerator: 4 to 21 MV photon beams, maximum dose at higher depth with energy, no radioactive source, radiation only when the source is switched is ON, uniform dose absorption, 1mm source - nearly point source, small Penumbra, electron beam of various energies possible
Telecobalt: 1.25 MeV γ Photon, maximum dose at depth of 5 mm, source to be changed every 4 to 5 years, leakage radiation present even while the beam is off, relatively uniform distribution, 1-2 cm source diameter, larger penumbra, gamma Photon only