  • 3 commonly used radiotherapy technologies
    • External beam radiation therapy (EBRT or XRT or teletherapy)
    • Brachytherapy or sealed source radiation therapy
    • Systemic radioisotope therapy or unsealed source radiotherapy
  • Radioactivity
    A process whereby a nucleus that contains an excess energy undergoes a transformation to a more stable state by emitting energy in the form of particles or electromagnetic radiation
  • Radioactive half-life
    The time interval in which the original activity level, tabulated as disintegrations per second (dps) decreases by one half
  • Radionuclide/Radioisotopes and their half-lives
    • Polonium-215 (0.0018 sec)
    • Bismuth-212 (60.5 sec)
    • Sodium-24 (15 hrs)
    • Iodine-131 (8.07 days)
    • Cobalt-60 (5.26 yrs)
    • Radium-226 (1,600 yrs)
    • Uranium-238 (4.5 billion yrs)
  • Superficial X-Ray tube (Philips RT 100)
    • X-ray produced at 50-150 kV
    • Varying thickness of filtration (usually 1-6 mm Al) are added to harden the beam
    • Superficial treatments are usually given with the help of applicators or cones
    • SSD range 15 to 20 cm
    • Dose is highly dependent on source-skin distance, filtration and applicator area
    • Usually operated at 5-8 mA
    • Useful for irradiating tumor confined to about 5 mm depth
  • Issues with superficial radiotherapy
    • Short focus to skin distance (FSD) and hence high output and large influence of inverse square law
    • Calibration difficult due to strong dose gradient i.e. dose fall off and electron contamination
  • Low-energy machines
    • Superficial Equipment
    • Orthovoltage Units
  • Grenz rays
    • 10-15 kVp
    • Treatment of inflammatory disorders (Langerhans' cells), Bowen's disease, patchy stage mycosis fungoides, herpes simplex
  • Contact therapy
    • Superficial skin lesions
    • Endocavitary treatments for curative intent (rectal)
  • Superficial equipment
    • 50-150 kVp
    • Skin cancer and tumors no deeper than 0.5 cm
  • Orthovoltage machines
    • 150-500 kVp
    • Skin, mouth, and cervical carcinoma
    • Experience limitation in the treatment of lesions deeper than 2 to 3 cm
  • Orthovoltage units
    • Uses "conventional" X-Ray tube with electrons accelerated by an electric field
    • Stationary anode (in contrast to diagnostic tubes which have a rotating anode)
    • Filtration important
  • Limitations of low energy machines
    • Can not reach deep-seated tumors with an adequate dosage of radiation
    • Do not spare skin and normal tissues
  • Differences between superficial and orthovoltage
    • Superficial: 50 to 150kVp, small skin lesions, maximum applicator size typically < 7cm, typical FSD < 30cm, beam quality measured in HVL aluminium (0.5 to 8mm)
    • Orthovoltage: 150 to 500kVp, skin lesions, bone metastases, applicators or diaphragm, FSD 30 to 60cm, beam quality in HVL copper (0.2 to 5mm)
  • Orthovoltage units

    • Uses conventional X-ray tube
    • Energy range 150- 500 kV X-rays
    • Mostly used around 250 - 300 kVp
    • Applicators are used in superficial therapy
    • Treatment depths of around 20 mm
    • Penetration sufficient for palliative treatment of bone lesions relatively close to the surface (ribs, spinal cord)
  • Disadvantages of deep X-ray
    • Higher dose to bone - photoelectric absorption
    • Maximum dose on the surface hence higher skin dose
    • Treatment to a depth of only a few centimeters possible
    • High scattered radiation and larger penumbra
  • Orthovoltage X-ray filters
    • Copper and sometimes Tin in addition to Aluminum
    • Aluminum is placed distal to copper to remove soft secondary radiation
    • Copper is placed distal to Tin to remove soft secondary radiation when beam interacts with Tin
  • Orthovoltage X-ray operation
    • Usually operates at SSD (Source to surface distance) 50 - 70 cm with or without cone
    • Has moveable lead shields
    • Low dose rate due to long SSD and heavy filtration of beam
  • Superficial machine or kilovoltage units
    • Operating with X-rays tubes at an accelerating potential < 50 kV
    • Superficial therapy refers to potentials of 50-150
    • Low/poor penetrating ability
    • Additional filters in the form of aluminum, tin and copper are added for beam hardening
    • Used in treatment of superficial lesions
    • Cone is directed to skin surface of patient
    • Lead shielding to be placed directly to shield surrounding areas
    • Treatment distance is usually 15 - 20 cm to decrease depth dose
    • Higher patient dose due to distance
    • Backscatter is high due to low energy and increases with increased field distance
  • Orthovoltage therapy or deep therapy
    • Energy: 200 - 300 kV
    • Tube current: 10 - 20 mA
    • HVLs: 1 - 4 mm Cu
    • Cones or movable diaphragm (continuous adjustable field size)
    • SSD: 50 cm
    • Application: tumor located < 2-3 cm in depth
  • Limitations of orthovoltage therapy
    • Skin dose
    • Depth dose distribution
    • Increase absorbed dose in bone
    • Increase scattering
  • Contact X-ray brachytherapy
    • Also called "CXB", "electronic brachytherapy" or the "Papillon Technique"
    • A type of radiation therapy using X-rays applied close to the tumour to treat rectal cancer
  • Megavoltage
    A clinical modality consisting of the administration of high energy (1 megavolt or greater) to a deep-seated cancer (e.g., prostate or brain cancer) or cerebrovascular defects by an MRT unit (e.g., linear accelerator or 60Co unit)
  • Radioisotope therapy
    • Delivers radiation directly into the cancer cells, usually as a capsule, drink or injection into a vein
    • Cancer cells absorb the radioactive substance more than normal cells, receiving a higher dose of radiation and causing the cells to die
  • Betatron
    • A cyclic accelerator that produces high-energy electrons for radiotherapy
    • The magnetic field of the betatron deflects electrons into a circular orbit, and an increasing magnetic orbital flux produces an induced circumferential electric field that accelerates them
    • Developed by KERST in 1941
    • Replaced by LINAC in 1950
  • Disadvantages of betatron
    • Low dose
    • Limited field size
    • Needs large treatment room due to large size
    • Limited motion
  • Van de Graaff generator
    • An electrostatic generator which uses a moving belt to accumulate electric charge on a hollow metal globe on the top of an insulated column, creating very high electric potentials
    • Produces very high voltage direct current (DC) electricity at low current levels
    • 1931 (MIT)
    • 40 feet high Electrostatic device capable of operating at 5,000,000 volts
    • 2 MeV Clinical Van de Graaff X-ray machine
    • Invented by Robert van de Graaff
  • Machines using isotopes
    • Teletherapy - Treatment in which source of radiation is at some distance from the patient
    • Brachytherapy - The source of radiation is very close to treated tissue
  • Cobalt-60
    • On and off shield source
    • Emits two photons/disintegration: 1.17 and 1.33 Mev which is useful in radiation therapy
    • The dose rate is constantly decreasing and adjustment of treatment is done periodically
    • Large size: penumbra
    • LINAC: small focal spot
    • Cobalt: near: small source: activity and dose is reduced
  • Cesium-137
    • Teletherapy isotope machine
    • Same with the shield source of Cobalt-60
    • 20 to 30 cm with 0.662 MeV
  • Radioactive sources (γ ray equipment)
    • Cobalt 60
    • Cesium 137
  • Telecurie units
    • Cs-131
    • Cobalt-60 Unit
  • Features of a teletherapy source
    • High energy gamma ray emission
    • High specific air kerma rate constant
    • Simple means of production
    • Low cost
    • High specific activity
    • Long half-life
  • Cesium-137
    • Photon energy 0.66MeV
    • Relatively large source to relatively low specific activity
    • Medium FSD (around 60cm)
    • No isocentric mounting - similar to orthovoltage equipment in set-up
    • Not sold anymore and should not be in use
  • Natural cobalt (59Co)

    • Kobald, from the German for goblin or evil spirit
    • Discovered in 1735
    • Brittle hard metal similar to iron and nickel
    • Found in minerals and meteorites
    • Salts and glass oxides are deep blue in colour
  • Differences between orthovoltage and telecobalt unit
    • Orthovoltage: 150-500 KV x-rays, maximum dose on the skin, treatment to a depth of few centimeters, higher absorption by bone, non uniform dose distribution, higher side scatter hence larger penumbra, mostly isocentric unit
    • Telecobalt: 1.25 MeV γ Photon, maximum dose at depth of 5 mm, relatively uniform dose absorption, higher penetration deep seated tumours, more of forward scatter, lesser penumbra
  • Differences between linear accelerator and telecobalt unit
    • Linear Accelerator: 4 to 21 MV photon beams, maximum dose at higher depth with energy, no radioactive source, radiation only when the source is switched is ON, uniform dose absorption, 1mm source - nearly point source, small Penumbra, electron beam of various energies possible
    • Telecobalt: 1.25 MeV γ Photon, maximum dose at depth of 5 mm, source to be changed every 4 to 5 years, leakage radiation present even while the beam is off, relatively uniform distribution, 1-2 cm source diameter, larger penumbra, gamma Photon only
  • Telecobalt unit
    • Introduced in the 1950's, being replaced by linacs
    • The first practical radiation therapy treatment unit to provide a significant dose below the skin surface and simultaneously spare the skin
    • Still used in developing countries: simpler design, cost, little tech support
  • Properties of telecobalt unit
    • Photon energy around 1.25MeV
    • Specific activity large enough for FSD of 80cm or even 100cm
    • Therefore, isocentric set-up possible
    • Constantly emit radiation
    • 60Co source must be shielded in a protective housing (source head)
    • Source head is a steel shell filled with lead (may be up to 2 ft in diameter)
  • Application of telecobalt unit
    • To treat cancers of the head and neck area, breast, spine, and extremities
    • Areas just below the skin surface