External beam radiation therapy (EBRT or XRT or teletherapy)
Brachytherapy or sealed source radiation therapy
Systemic radioisotope therapy or unsealed source radiotherapy
Radioactivity
A process whereby a nucleus that contains an excess energy undergoes a transformation to a more stable state by emitting energy in the form of particles or electromagnetic radiation
Radioactive half-life
The time interval in which the original activity level, tabulated as disintegrations per second (dps) decreases by one half
Radionuclide/Radioisotopes and their half-lives
Polonium-215 (0.0018 sec)
Bismuth-212 (60.5 sec)
Sodium-24 (15 hrs)
Iodine-131 (8.07 days)
Cobalt-60 (5.26 yrs)
Radium-226 (1,600 yrs)
Uranium-238 (4.5 billion yrs)
Superficial X-Ray tube (Philips RT 100)
X-ray produced at 50-150 kV
Varying thickness of filtration (usually 1-6 mm Al) are added to harden the beam
Superficial treatments are usually given with the help of applicators or cones
SSD range 15 to 20 cm
Dose is highly dependent on source-skin distance, filtration and applicator area
Usually operated at 5-8 mA
Useful for irradiating tumor confined to about 5 mm depth
Issues with superficial radiotherapy
Short focus to skindistance (FSD) and hence high output and large influence of inverse square law
Calibration difficult due to strong dose gradient i.e. dose fall off and electron contamination
Endocavitary treatments for curative intent (rectal)
Superficial equipment
50-150 kVp
Skin cancer and tumors no deeper than 0.5 cm
Orthovoltage machines
150-500 kVp
Skin, mouth, and cervicalcarcinoma
Experience limitation in the treatment of lesions deeper than 2 to 3 cm
Orthovoltage units
Uses "conventional" X-Ray tube with electrons accelerated by an electric field
Stationary anode (in contrast to diagnostic tubes which have a rotating anode)
Filtration important
Limitations of low energy machines
Can not reach deep-seated tumors with an adequate dosage of radiation
Do not spare skin and normal tissues
Differences between superficial and orthovoltage
Superficial: 50 to 150kVp,small skin lesions, maximum applicator size typically < 7cm, typical FSD < 30cm, beam quality measured in HVL aluminium (0.5 to 8mm)
Orthovoltage: 150 to 500kVp, skin lesions, bone metastases, applicators or diaphragm, FSD 30 to 60cm, beam quality in HVL copper (0.2 to 5mm)
Orthovoltage units
Uses conventional X-ray tube
Energy range 150- 500 kV X-rays
Mostly used around 250 - 300 kVp
Applicators are used in superficial therapy
Treatment depths of around 20 mm
Penetration sufficient for palliative treatment of bone lesions relatively close to the surface (ribs, spinal cord)
Disadvantages of deep X-ray
Higher dose to bone - photoelectric absorption
Maximum dose on the surface hence higher skin dose
Treatment to a depth of only a few centimeters possible
High scattered radiation and larger penumbra
Orthovoltage X-ray filters
Copper and sometimes Tin in addition to Aluminum
Aluminum is placed distal to copper to remove soft secondary radiation
Copper is placed distal to Tin to remove soft secondary radiation when beam interacts with Tin
Orthovoltage X-ray operation
Usually operates at SSD (Source to surface distance) 50 - 70 cm with or without cone
Has moveable lead shields
Low dose rate due to long SSD and heavy filtration of beam
Superficial machine or kilovoltage units
Operating with X-rays tubes at an accelerating potential < 50 kV
Superficial therapy refers to potentials of 50-150
Low/poor penetrating ability
Additional filters in the form of aluminum, tin and copper are added for beam hardening
Used in treatment of superficial lesions
Cone is directed to skin surface of patient
Lead shielding to be placed directly to shield surrounding areas
Treatment distance is usually 15 - 20 cm to decrease depth dose
Higher patient dose due to distance
Backscatter is high due to low energy and increases with increased field distance
Orthovoltage therapy or deep therapy
Energy: 200 - 300 kV
Tube current: 10 - 20 mA
HVLs: 1 - 4 mm Cu
Cones or movable diaphragm (continuous adjustable field size)
SSD: 50 cm
Application: tumor located < 2-3 cm in depth
Limitations of orthovoltage therapy
Skindose
Depthdose distribution
Increaseabsorbeddose in bone
Increasescattering
Contact X-ray brachytherapy
Also called "CXB", "electronic brachytherapy" or the "Papillon Technique"
A type of radiation therapy using X-rays applied close to the tumour to treat rectal cancer
Megavoltage
A clinical modality consisting of the administration of high energy (1 megavolt or greater) to a deep-seated cancer (e.g., prostate or brain cancer) or cerebrovascular defects by an MRT unit (e.g., linear accelerator or 60Co unit)
Radioisotope therapy
Delivers radiation directly into the cancer cells, usually as a capsule, drink or injection into a vein
Cancer cells absorb the radioactive substance more than normal cells, receiving a higher dose of radiation and causing the cells to die
Betatron
A cyclic accelerator that produces high-energy electrons for radiotherapy
The magnetic field of the betatron deflects electrons into a circular orbit, and an increasing magnetic orbital flux produces an induced circumferential electric field that accelerates them
Developed by KERST in 1941
Replaced by LINAC in 1950
Disadvantages of betatron
Lowdose
Limitedfieldsize
Needs largetreatmentroom due to large size
Limitedmotion
VandeGraaffgenerator
An electrostatic generator which uses a moving belt to accumulate electric charge on a hollow metal globe on the top of an insulated column, creating very high electric potentials
Produces very high voltage direct current (DC) electricity at low current levels
1931 (MIT)
40 feet high Electrostatic device capable of operating at 5,000,000 volts
2 MeV Clinical Van de Graaff X-ray machine
Invented by Robert vandeGraaff
Machines using isotopes
Teletherapy - Treatment in which source of radiation is at some distance from the patient
Brachytherapy - The source of radiation is very close to treated tissue
Cobalt-60
On and off shield source
Emits two photons/disintegration: 1.17 and 1.33 Mev which is useful in radiation therapy
The dose rate is constantly decreasing and adjustment of treatment is done periodically
Large size: penumbra
LINAC: small focal spot
Cobalt: near: small source: activity and dose is reduced
Cesium-137
Teletherapy isotope machine
Same with the shield source of Cobalt-60
20 to 30 cm with 0.662 MeV
Radioactive sources (γ ray equipment)
Cobalt 60
Cesium 137
Telecurie units
Cs-131
Cobalt-60 Unit
Features of a teletherapy source
High energy gamma ray emission
High specific air kerma rate constant
Simple means of production
Low cost
High specific activity
Long half-life
Cesium-137
Photon energy 0.66MeV
Relatively large source to relatively low specific activity
Medium FSD (around 60cm)
No isocentric mounting - similar to orthovoltage equipment in set-up
Not sold anymore and should not be in use
Natural cobalt (59Co)
Kobald, from the German for goblin or evil spirit
Discovered in 1735
Brittle hard metal similar to iron and nickel
Found in minerals and meteorites
Salts and glass oxides are deep blue in colour
Differences between orthovoltage and telecobalt unit
Orthovoltage: 150-500 KV x-rays, maximum dose on the skin, treatment to a depth of few centimeters, higher absorption by bone, non uniform dose distribution, higher side scatter hence larger penumbra, mostly isocentric unit
Telecobalt: 1.25 MeV γ Photon, maximum dose at depth of 5 mm, relatively uniform dose absorption, higher penetration deep seated tumours, more of forward scatter, lesser penumbra
Differences between linear accelerator and telecobalt unit
Linear Accelerator: 4 to 21 MV photon beams, maximum dose at higher depth with energy, no radioactive source, radiation only when the source is switched is ON, uniform dose absorption, 1mm source - nearly point source, small Penumbra, electron beam of various energies possible
Telecobalt: 1.25 MeV γ Photon, maximum dose at depth of 5 mm, source to be changed every 4 to 5 years, leakage radiation present even while the beam is off, relatively uniform distribution, 1-2 cm source diameter, larger penumbra, gamma Photon only
Telecobalt unit
Introduced in the 1950's, being replaced by linacs
The first practical radiation therapy treatment unit to provide a significant dose below the skin surface and simultaneously spare the skin
Still used in developing countries: simpler design, cost, little tech support
Properties of telecobalt unit
Photon energy around 1.25MeV
Specific activity large enough for FSD of 80cm or even 100cm
Therefore, isocentric set-up possible
Constantly emit radiation
60Co source must be shielded in a protective housing (source head)
Source head is a steel shell filled with lead (may be up to 2 ft in diameter)
Application of telecobalt unit
To treat cancers of the head and neck area, breast, spine, and extremities