  • Components in HPLC
    • Solvent (MP) reservoir
    • Pump & solvent mixer
    • Injection system
    • Column
    • Detector
  • MP reservoir
    • Solvents used as MP must be of high purity
    • Degassed (bubbles & microbubbles are removed)
    • Filtered through 0.45 µm or 0.2 µm filter membrane
  • Isocratic elution
    Fixed MP composition throughout the analysis
  • Gradient elution
    MP composition changes at certain points during the analysis
  • Gradient elution
    • Allows adequate separation of early eluting peaks without the later eluting peaks becoming too dispersed
    • Starts with a weaker eluting strength MP to allow least retained analytes to have more retention in the SP
    • The eluting strength of the MP increases with time to elute strongly retained analytes earlier
  • Pump
    • Drives mobile phase through the entire LC system
    • Has an in-built mixing chamber for homogenizing aqueous and organic solvents
  • Sample Injection system
    • Can either be manual injection loop or autosampler
  • Column
    • Located within a heating element controlled by a thermostat to ensure a constant column temperature
    • Typical column is 5-30 cm in length & 3-4.6 mm in inner diameter
  • Detectors used in HPLC
    • UV-VIS
    • Evaporative Light Scattering Detector (ELSD)
  • UV-VIS detector
    • Most widely used
    • Quantification based on absorption of UV-VIS radiation
    • Variable wavelength UV-VIS detector can measure across the entire wavelength range
    • Can be used to measure analytes that absorb UV-VIS radiation
    • Wide linear dynamic range
    • Non-destructive
    • Inexpensive
    • Robust
  • Evaporative Light Scattering Detector
    • Based on the light scattering properties of the analyte
    • Concentration of the analyte is linearly related to the amount of light scattered
    • Universal detector
    • Can be used for analytes that DO NOT absorb UV-VIS radiation
    • Limited linear dynamic range
    • Lower sensitivity than UV-VIS
    • MP must be volatile
  • Common problems in HPLC
    • Pressure (abnormally high or low)
    • Baseline (noisy or drifting)
    • Peak shape (peak broadening or split peaks)
    • Retention (shifting retention time)
  • Troubleshooting abnormally high pressure
    1. Plugged inlet frit
    2. Column blockage
    3. Precipitation along the solvent lines
  • Troubleshooting abnormally low pressure
    1. Presence of air bubbles between the pump and MP reservoir
    2. Leaking pump valve / seals
  • Troubleshooting noisy baseline
    1. Detector lamp failing
    2. Contaminated detector cell
    3. Air bubble in the system / detector
  • Troubleshooting drifting baseline
    1. Contaminated MP
    2. Temperature fluctuations
    3. Column bleeding
  • Troubleshooting peak broadening
    1. Sample solvent strength is higher than MP
    2. Injection volume too large (column overload)
    3. Low MP flow rate
  • Troubleshooting split peaks
    1. Partially blocked inlet frit
    2. Small void at the column inlet
    3. Poor separation of two peaks
  • Troubleshooting shifting retention time
    1. Column aging / contamination
    2. Change in MP composition
    3. Presence of air in pump
  • Separation principles in TLC
    • Stationary phase (silica plate) is spread in a thin layer on the surface of a plate
    • Analyte prepared in a volatile solvent, are placed as spots
    • Analyte carried by the mobile phase (solvent) which travels up the stationary phase by capillary action
    • Separation is based on the relative affinity of the analyte for the stationary phase vs the mobile phase
    • More polar analytes adsorbed with higher affinity to the silica gel travel slower, decreasing the retardation factor (Rf)
  • Stationary phase in TLC
    • Silica gel or silica-gel based materials
    • Can be modified to make it non-polar
  • Mobile phase in TLC
    • Volatile liquids, e.g. mixture of chloroform/methanol/water (180:15:1 v/v)