collimated X-Ray beam; X-Ray tube rotates around the patient (through the gantry)
In computed tomography, narrow x-ray beam are aimed at the patient and quickly rotates around the body of the patient producing signals and processed by the machine to produce cross-sectional images (tomograms)
early sensors for computed tomography were scintillation detectors with photomultiplier tubes excited by sodium iodide (NaI) crystals
modern detectors are solid state scintillators coupled to photodiodes or are filled with low-pressure xenon gas
An image is obtained in CT by computer processing of the digital readings of the detectors.
motorized x-ray tube that rotates around the circular opening (gantry) together with collimator
patient lies on the patient couch that slowly move through the gantry
while x-ray tube rotates, it shoots narrow beam/collimated beam throughout the body
ct scanner uses specialized detectors (scintillators with PMT), which is directed opposite to the x-ray tube
detectors pick-up x-rays that pass through the patient
in one full rotation, CT generates 2D images
image processing
3D
image attributes of computed tomography: thin transverse sections of the body are acquired representing an absorption pattern or X-ray attenuation of each tissue
Absorption values are expressed as Hounsfield Units (HU)
CT number for dense bone: 3000
CT number for muscle: 50
CT nmber for white matter: 45
CT number for gray matter: 40
CT number for blood: 20
CT number for cerebrospinal fluid: 15
CT number for water: 0
CT number for fat: -100
CT number lungs: -200
CT number for air: -1000
computed tomography has good contrast resolution allowing differentiation of tissues with similar physical densities
tomographic acquisition eliminates the superposition of images of overlapping structures
advanced scanners can produce images that can be viewed in multiple planes or as volumes. Any region of the body can be scanned
computed tomography has become diagnostic modality of choice for a large number of disease entities
computed tomography is useful for tumor staging
computed tomography has high cost of equipment and procedure
computed tomography has high dose of ionizing radiation per examination
CT produces artifacts from high contrast objects in the body such as bone or devices
table increment/pitch ratio: movement of patient couch
slow pitch ratio = higher patient dose, good contrast resolution
fast pitch ratio = lower patient dose, poor contrast resolution