A goal of the urinaysis laboratory is to maximizeproductivity and testingquality, while keeping costs and turnaround time at a minimum
Studies have shown that the major variable in urinalysis testing is the conscientiousness of the laboratory personnel in their timing and interpretations of the color reactions.
Automation brings qualitycontrol,precision, fasterTAT (turn-around-time), and standardizationofresults to a method that previously was uncontrolled
Urine chemistry Semiautomated analyzer
Daily maintenance is minimal and includes cleaning the reagent strip platform and emptying the reagent strip waste container.
The semiautomated instrument requires the operator to:
Dip the reagent strip into a urine sample that has been mixed well.
Blot the strip to remove excess urine.
Place the strip onto the reagent strip platform.
Press the analyze/enter button.
REFLECTANCE PHOTOMETRY
Reflectance photometry uses the principle that lightreflection from the test pads decreases in proportion to the intensity of the color produced by the concentration of the testsubstance
A monochromatic light source is directed toward the reagentpads by placing a filter between the light source and the reflective surface of the pad or by using a light-emitting diode (LED) to provide the specific wavelength needed for each test pad color reaction
REFLECTANCE PHOTOMETRY
The light is reflected to a photodetector, as well as a converter that is either analog or digital. The instruments compare the amount of light reflection with that of knownconcentrations and then display or print concentration units or transmit data to a laboratory information system (LIS).
Reflectance photometry quantifies the intensity of the colored product produced on the reagent strip reaction pads.
REFLECTANCE PHOTOMETRY
When light strikes a matte or unpolished surface (e.g., a reagent strip), some light is absorbed, and the remaining light is scattered or reflected in alldirections. The scattered light is known as diffuse reflectance.
Photometers are calibrated using reflectancestandards such as magnesium carbonate or barium sulfate that "completely" reflect all incident light
The relationship between reflectance and concentration is not linear (inversely proportional)
Automated Microscopy Analyzers
Decrease laborcosts and increasing productivity in the urinalysis laboratory
Uncentrifuged urine is used, the time spent in handling and preparing concentrated urine sediment for manual microscopy is eliminated
Increased standardization of the microscopic examination, which enhances the accuracy and reproducibility (precision) of results
It is an automated system that performs the microscopic examination of urine, as well as cell counts on body fluids
Uses patented technologies to capture and automatically classify digital images of urineparticles.
SOFTWARE USED: AUTO PARTICLE RECOGNITION SOFTWARE or Proprietary neural network software
Iris iQ 200
APR Pre-classifies urine particles in the photographs based on size, shape, texture, and contrast in to 1 to 12 categories. Using the computer monitor, the user can reviewresults,visually assess the particles present, and subclassify them into the 26 or 27 additional categories
Iris iQ 200
The field of view of the microscope is coupled to a digital video camera, and stroboscopic illumination freezes the particles in motion as they stream past, which ensures blur-free imaging. With each sample, the camera captures 500 frames, digitizes them, and sends them to a computer for processing
iQ200 Auto-classification and Subclassification Categories for Urine Sediment Particles
12auto-classifiedsediments are primary parameters measured
Blood cells
Autoclassified by Analyzer:
RBC
WBC
WBC clump
Subclassified by User:
RBC clump
iQ200 Auto-classification and Subclassification Categories for Urine Sediment Particles
Crystals
Autoclassified by Analyzer: unclassifiedcrystals
Subclassified by User:
Amorphous
Calciumcarbonate
Calciumoxalate
Calciumphosphate
Triple phosphate
Uric acid
Cystine
Tyrosine
Leucine
Unclassified crystals
iQ200 Auto-classification and Subclassification Categories for Urine Sediment Particles
Casts
Autoclassified by Analyzer:
Hyaline
Unclassified casts
Subclassified by User:
Granular
Cellular
Waxy
Broad
RBCs
WBCs
EpithelialCells
Fatty
Unclassifiedcasts
Epithelial cells
Autoclassified by Analyzer:
Squamous
Non squamous
Subclassified by User:
Transitional
Renal
iQ200 Auto-classification and Subclassification Categories for Urine Sediment Particles
Principle: Cell flow cytometry (With Electricalimpedance for newer models)
Urine particles are identified and categorized by fluorescent staining characteristic, light scatter, electrical impedance, and adaptive cluster analysis
To perform an automated urine sediment analysis, 1.2ml of uncentrifugedurine is aspirated into the instrument and divided into two channels: the sedimentchannel for urineparticleanalysis and the bacteriachannel for bacteriastaining and detection
Each channel has a specific stain that targets surface and/or internalcomponents of the cells. The stained urine sample passes through the flowcell, where it is hydrodynamically focused and presented to a red semiconductor laser (635m)
Particles are identified by measuring the height and width of the fluorescent and lightscattersignals, which are presented in scattergrams and histograms
In the bacteria channel, the diluent stabilizes the pH and lyses the nonbacterial particles, reducing interference from amorphous crystals. The stain is specific to the ribonucleic acid (RNA) in a bacterial cell, eliminating any nonspecific staining of debris
The width of the fluorescent signal measures cellular inclusions, and the width of forward light scatter measures the lenath of cells
A latex particle quality control system monitors performance, and quality control records can be viewed on the instrument screen in a Lew-Jennings graph. The analyzers can store up to 10,000 patient results, including scattergrams, histograms, and specimen characteristics, which can be saved and viewed by the user
For automated particle analysis, the UF-10001 analyzer requires a 4 mL sample volume; however, if the instrument is used in the manual mode, only 1 mL of urine is required Changes in the UF-10001 analyzer include a separate channel for bacterial analysis and the monitoring of lateral or side scatter, which improves detection of bacteria
Sysmex UF-Analyzers
Bacteria is detected specifically by the side light scatter
Flagged Particles: Non hyaline (pathologic) cast, Crystals, small round cells (transitional or renal cells), Yeast, Mucus, Sperm
Additional notes:
Determining the specific identity of elements in "flagged" specimens requires a manual microscopic review of the urine by the user.
Flaggedparticles = Particles that cannot be auto-enumerated by the machine and mustbe confirmed by the lab personnel thru manual microscopy
The newest Sysmex model, the UN-2000, is an automated, modular, and scalable urinalysis system that combines urine flow cytometry with digital image analysis
Additional notes:
The UF-5000 utilizes fluorescent flow cytometry coupled with a blue semiconductor laser to measure particle length, particle volume, particle internal complexity, and total amount of nucleic acid contained within the particle
Additional notes:
The Sysmex UD-10 is a complementary digital imaging system designed to capture detailed images of urine particles in urine samples that require further review. All captured urine particle images are classified, based on size, into eight different classes. Then the technologist has the opportunity to review the captured images and identify the type(s) of urine particles present
Additional notes:
The UN-2000 uses a single-screen data management system for monitoringanalyzerstatus, result reporting, and reflex rule setting, as well as providing the connection point for interface with the LIS or middleware system
Full Automated Urinalysis system
Performing fullyautomated urinalyses requires combining a urine chemistry analyzer with a microscopy analyzer.