Cards (18)

  • Ultrasound is
    • Longitudinal
    • Frequency greater then 20kHz
    • An ultrasound transducer is used to emit pulses of ultrasound into the body
    • Pulses are reflected by boundaries between different tissues
    • Transducer detects reflected pulse then reflected intensity and delay time can be used to give information about type of tissue and depth.
    • A computer takes this data to produce an image
  • A-Scan
    • Ultrasound pulses in one direction
    • Used to find depth or distance by producing a trace eg thickness of skin or muscle
  • B-Scan
    • Ultrasound pulses in many directions with multiple detectors
    • Produces a 2d or 3d image eg pre-natal scanning
  • Ultrasound transducer uses the piezoelectric effect to emit and detect ultrasound
  • Piezoelectric effect - When a pd is applied to a material it causes compression or extension
  • Emission of Ultrasound
    • Piezoelectric crystal is connected to alternating pd
    • Causing crystal to vibrate
    • Emitting ultrasound
  • Detection of Ultrasound - Piezoelectric can happen in reverse - compession or extension causes a pd to be induced
    • Ultrasound is absorbed by crystal
    • Causing cystal to vibrate
    • Induces an alternating pd
    • This is send to a computer to produce a trace (A) or image (B)
  • The depth of a tissue can be determined using s=vt
    • s - distance from emitter to tissue to detector (depth is s/2)
    • v = speed of sound in the tissue
    • t = time for the ultrasound pulse to reach the tissue and be reflected back
  • Acoustic inpedance is a property of a material defined as the density of the material multiplied by the speed of sound through it
  • The greater the differance in acoustic impedances of two materials the greater the intensity of the reflected ultrasound at the boundary between the materials
  • The fraction of ultrasound reflected is the differance between the acoustic impedances squared divided by the sum of the acoustic impedances squared
  • Transmitted intensity is 1-reflected intensity
  • Impedance matching - if there is any air between the transducer and skin most of the ultrasound will be reflected and little will be transmitted into the body
  • An imedance matched gel is applied to the skin
    • The acoustic impedance of the gel and skin are similar
    • So very little ultrasound is reflected by the skin at the boundary
    • So most is transmitted into the body
  • The resolution of a scan (smallest distinguishable feature) is equal to the wavelength of the ultrasound. Shorter wavelengths give finer detail but only last shorter depths
  • Dopler effect - A change in frequency/wavelength due to relative motion of source and observer
  • Measuring the flow of blood in arteries using ultrasound
    • Transducer placed at angle to artery and emits pulse
    • Pulse is reflected by moving blood
    • Transducer detects reflected pulse which has changed frequency and wavelength
    • The change in frequency is related to speed of blood