Kesson (1978) provided evidence that chamber counts on centrifuged urine sediments are more reliable in predicting renal functional abnormalities than is a conventional method using cells per HPF
If obtaining a 12 mL specimen is not possible: Specimen volume noted on report, For correction: Example: if only 6 mL of urine is centrifuged, results are multiplied by 2
Routine Urine Analysis - Examination of the Sediment
Volume of sediment examined: Glass slide method: 20uL covered by a 22 x 22 mm cover slip, Commercial systems: control volume of sediment by providing slides with chambers
Urine sediment examined in a minimum of 10 LPO and 10 HPO, LPO scanning of cover-slip perimeter is recommended for detection of casts
Use reduced light when examining unstained sediment by bright-field microscope
Slides prepared by cytocentrifugation and stained with Papanicolaou stain
Performed by the cytology laboratory
Purpose: For detection of malignancies of the lower urinary tract, To provide definitive information about renal tubular changes associated with transplant rejection; viral, fungal, and parasitic infections; cellular inclusions; pathologic casts; and inflammatory conditions
Light source (Lamps/bulbs: Tungsten, Halogen, Fluorescent, LED, Regulated by a Rheostat, Filters providing enhanced contrast and color correction)
Condenser (gather the light coming from the light source and to concentrate that light in a collection of parallel beams into the specimen)
Field and iris (aperture) diaphragm (Iris Diaphragm controls the amount of light passing through the slide or specimen, Field Diaphragm may also be present in the light source; may be opened or closed; used in focusing the light which passes up through the condenser)
Objectives (Perform the initial magnification, Inscriptions: Objective magnification, Total magnification, thickness of the cover glass, labeled OIL or HI or OEL, numerical aperture)
Oculars/Eyepiece lens (Responsible for further resolution)
1. Adjustments done to provide optimal viewing of illuminated objects
2. Steps: Place a slide on the stage and focus using LPO with the condensers raised, Close the field diaphragm, Lower he condenser until the edges of the field diaphragm are sharply focused, Center the image of the field diaphragm with the condenser centering screws, Open the field diaphragm until its image is at the edge of the field, Remove an eyepiece and look down through the eyepiece tube, Adjust the aperture diaphragm until approximately 75% of the field is visible, Replace the eyepiece
The use of polarized light aids in identification of birefringent elements: cholesterol on OFB, fatty casts, and crystals
Birefringent – a property indicating that the element can refract light in 2 dimensions at 90 degrees to each other
Obtaining a polarized light: Brightfield microscopes could be used, 2 filters installed in a cross formation: 1st filter - Polarizing filter, 2nd filter - Analyzer
Interpretation: Nonbirefringent objects = no refracted light = no light reaches analyzer = object appears black, Birefringent objects = refracted light = refracted light reaches analyzer = object appears white/colored against a black background
Produces a 3-dimensional microscopy-image and layer-by-layer imaging of a specimen
Split the light ray so that the beams pass through different areas of the specimen. The light interference produced by the varied depths of the specimen generates a 3-dimensional image
Used to enhance visualization of specimens not viewed with a bright-field microscope
Aids in identification of Treponema pallidum
A bright-field microscope is adapted, Replacing the condenser with a dark-field condenser that contains an opaque disk, Disk blocks light from directly entering the objective, Light rays pass around the specimen and are scattered, producing a bright image on a dark background