Fast, accurate microbial identification, and antibiotic susceptibility testing
Difference with BACTEC and BACT/ALERT: VITEK 2 has the antibiotic susceptibility option.
Fully automated system that performs bacterial identification by biochemical analysis using colorimetry.
Imagine using an API method but automated.
We wanted automation to limit human interventions and human errors.
It focuses on the clinical microbiology laboratory and provides increased levels of automation and capacity for higher volume laboratories.
Rampant in tertiary laboratories
VITEK® 2 (AUTOMATION)
ITEK 2 system can accurately identify some bacterial strains in as little as twohours.
In contrast, BACT/ALERT which can only indicate whether the result is positive or negative.
In addition to being able to identify bacteria, the VITEK 2 compact system is able to identify multiple species of yeast.
They also provide an option of automatic pipetting and dilution for antimicrobial susceptibility testing.
Advantages
Versatile
Eliminated the subjectivity of reading test reactions
May offer automatic pipetting and dilution for antimicrobial testing
Provides increased levels of automation and capacity for higher volume laboratories
Disadvantage
Some microorganisms that lead to correct MIC reports or yield unreliably (e.g., Pseudomonas)
VITEK 2 Cards
The reagent cards have 64 wells (64 tests) that can each contain an individual test substrate.
More detection of various metabolic activities than API
Substrates measure various metabolic activities such as acidification, alkalinization, enzyme hydrolysis, and growth in the presence of inhibitory substances.
Collab: BACTECTM and VITEK® 2 Workflow
BD BACTEC FX blood culture flags positive
Transfer 1 mL blood culture fluid to a 1.5 mL microcentrifuge tube.
Add 200 microliter (μl) of Lysis buffer.
Vortex for 10 seconds.
Centrifuge for 2 min at 13000 rpm.
Discard the supernatant and wash the pellet in 1 mL of wash buffer.
Centrifuge for 2 min at 13000 rpm.
Collab: BACTECTM and VITEK® 2 Workflow
7. Centrifuge for 2 min at 13000 rpm
8. Discard the supernatant then suspend the pellet in 500 microliters of sterile 0.45% normal saline and let it settle.
9. Transfer bacterial suspension to make 0.50 to 0.63 McFarland standard solution.
10. Set-up VITEK 2 as per manufacturer recommendations.
Can go directly; no need to prepare for culture plates or tubes.
VITEK 2 Compact || Microbial Identification || AST || Automation in Microbiology Used for microbial identification and antibiotic sensitivity
In the machine, identification occurs in various biochemical tests.
Antibiotic sensitivity is done by MIC (Minimum Inhibitory Concentration) method for three different concentrations for each drop. The sensitivity is detected by the colorimetric method.
After the preparation of microorganism suspension, reagent cards and the suspension in the test tube are placed into their designated slots in a cassette or special rack. The cassette is loaded in the chamber
Once the start button is pressed, it creates a negative pressure inside the filling chamber that sucks all the air inside the chamber, including the air from the tube and card.
Air returns into the chamber; causing the organic suspension to move through the transfer tube and into the micro-channels. This fills all the test wells (Batch 2025). The dehydrated reagents are dissolved in this suspension.
Every 15 minutes, each card undergoes scanning in the colorimetric scanner to detect color development that occurs due to biochemical reaction or sensitivity within the card.
After identification and sensitivity are finalized, the reagent card is pushed into the base tray.
Once the tray is full, all cards are discarded into the yellow bin.