Radioisotopes (cannot control, must dispose appropriately)
Scanning electron microscopes – electron beam excites sample to produce x-rays, often paired with XRF analysis
Synchrotrons – bright light sources are used for research and very sensitive XRF analysis
Positron and other particle beams – All high energy particle beans ionise materials so that they give off x-rays. SEM and PIXE are most common beams used.
The tube continuum – as you increase the accelerating voltage, the number of x-rays emitted increased, x-rays appear at lower wavelengths (higher intensity)
The line spectra – sharp intense beam at particular wavelengths that come off of the contimuum. It's made from the metals in the anion of the tube. When the incoming electron strikes the inner orbital electron with high force to displace it (produces an ion). Just at a shorter wavelength to elements interested in, which enhances them. Only produced when accelerating voltage exceeds the threshold voltage.
Most intense lines will be the K-series lines of each element in the sample as these are the most common transitions.
1. Dispersion of the x-rays from the sample into their individual wavelengths using a crystal followed by quantification
2. Each element emits x-rays at specific, unique wavelengths. To quantify the x-ray intensity, we need to disperse the spectrum
3. A crystal with a regular structure is used to disperse the system
4. Bragg's law states that when an x-ray beam is incident on a regular crystal substance, very strong diffraction of a given wavelength will occur when the wavelength strikes at an angle θ and the following relationship is followed: nλ = 2d x sin θ
5. Many crystals are usually required. Bragg's equation shows that crystals suitable for diffraction must have a lattice spacing (2d) of the same order of magnitude as the wavelength of incident x-rays to be diffracted.
6. By changing the angle of incidence, a different wavelength is diffracted
7. Most WD CRF are sequential measuring devices with a single detector, the crystal is rotated relative to the sample and the detector is rotated twice the speed of the crystal to maintain the angular relationship.
Simultaneous measurement of all emitted x-rays and electronic deconvolution of the pattern into individual lines (elements) by the detector based on energy
Uses a single energy dispersive detector capable of simultaneous measurements which makes it cheaper
As there are no crystals or goniometer the detector can be placed close to the sample meaning a higher intensity enters the detector which allows a lower power x-ray tubes to be used which is cheaper (not as much heat produced)
Combining it with modern electronics means it's a smaller instrument which is transferable
Auger effect - The primary x-ray beam excited a volume of the sample, a secondary x-ray from the sample is absorbed by an atom of a different element (and releases an x-ray), instead of escaping to the detector
Rayleigh scatter - primary x-rays are reflected by atoms and no fluorescence occurs. The amount of rayleigh scatter is proportional to the atomic number of the sample (high atomic number elements cause more scatter)
Compton scatter - primary x-rays strike the atom without causing fluorescence but with transfer of energy. The intensity of compton scatter bears a strong relationship to the mean atomic number of the sample
SLFP Standardless Fundamental Parameter - The algorithm computes both the intensity to concentration relationship and the absorption effects. Results typically within 10-20% of actual values
FP (with standards) NBS-GSC, NRLXRF, Uni-Quant ect.. - The concentration to intensity relationship is determined with standards, while the FP handles the absorption effects. Results typically within 5-10% of actual values.
XRF is a reference method, standards are required for quantitative results. Calibration plot is needed to compare spectral intensities of unknown samples to those of known standards.
There is no physical interaction between the sample and the instrument, so recalibration is only needed every 6 months to a year. Only a monitor is usually required to check calibration is still valid.