Review 1.2

    Cards (34)

    • Patient-Ventilator Synchrony
      Adequate ventilation and oxygenation, decrease WOB and patient comfort
    • Patient-Ventilator Dysynchrony
      Ineffective oxygenation and ventilation, increase WOB and patient discomfort
    • Increase in ventilators workload

      Can lead to ventilatory muscle fatigue and structural injury of the lung
    • It takes 24 hours to recover from ventilatory muscle fatigue
    • Dangers of Controlled Ventilation
      Ventilatory muscle weakness, deconditioning, atrophy
    • Things that can worsen patient-ventilator asynchrony
      • Pain
      • Anxiety
      • Procedural interactions
      • Other clinician-patient interactions
    • Types of patient-ventilator asynchrony
      • Trigger asynchrony
      • Flow asynchrony
      • Cycle asynchrony
      • Mode asynchrony
    • Trigger Asynchrony
      Vent is not triggering breath to patient even though pt demands it
    • Flow Asynchrony
      Poor coordination between the pts flow demand and the flow provided by the vent. Common in VC.
    • Cycle Asynchrony
      Poor coordination between ventilator and pt respiratory drive
    • Mode Asynchrony
      Asynchrony develops when the mode selected does not match the pts spontaneous vent efforts
    • Study the effects of patient-ventilator asynchrony (Slide 10)
    • Trigger
      Parameter on the ventilator that initiates a breath
    • Types of trigger
      • Time trigger
      • Flow trigger
      • Pressure trigger
    • Patient-triggered
      Pressure trigger or flow trigger
    • Controlled Breaths
      Time trigger
    • There would be no dips or jumps on none of the waveforms to identify a time trigger breath
    • When will a patient receive a time trigger breath
      Once the preset time interval is met. 60sec/RR=preset time interval
    • A jump at the end of exhalation will be shown on the flow-time scalar to identify a flow trigger breath
    • How the ventilator knows to deliver a flow trigger breath
      When a change in bias flow is detected
    • Flow trigger
      Usually set between 1 to 2 L/min
    • There is a dip at the beginning of the pressure-time scalar waveform to identify a pressure trigger breath
    • How the ventilator knows to deliver a pressure trigger breath
      Ventilator detects a pressure drop at the proximal airway due to inspiratory effort of the patient and vent begins trigger
    • Pressure Trigger
      Usually set between -0.5 to -1.5 cm
    • To decrease the patient's sensitivity (make it harder for the patient to initiate a breath)
      Increase the trigger
    • To increase the patient's sensitivity (make it easier for the patient to initiate a breath)
      Decrease the trigger
    • Trigger Work
      Portion of the WOB performed by the patient to trigger the ventilator to the inspiratory phase
    • How to correct autoPEEP
      Increase flow/ Decrease iT, Decrease RR or VT, increase peep, larger diameter ETT, provide bronchodilator, suction secretions
    • Types of trigger asynchrony
      • Missed trigger
      • Double trigger
      • Reverse trigger
      • Trigger delay
      • Auto triggering
    • Missed Triggering
      Inspiratory effort that does not trigger the vent. Fix By: correcting autoPEEP, adjusting trigger by decreasing trigger
    • Trigger Delay
      Occurs when there is increased time interval between diaphragm and ventilators initiation of breath. Fix by: decreasing trigger parameter (easier to breathe)
    • Double Triggering
      Most often in VC mode, set VT or IT is less than s neuro vent control center. Second breaths occurs immediately after term of og breath. Fix by: sedation, diff mode, increasing VT, Increasing IT to match neuro-iT
    • Reverse Triggering
      Occurs during controlled ventilation, ventilator initiates breath before diaphragm
    • Auto Triggering
      Ventilator initiates inspiration without corresponding patient effort due to inappropriate trigger sensitivity. Fix by: increasing vent trigger sensitivity, correct leaks, drain water