Inspiratory gas flow from the vent is unable to match the patients inspiratory flow demand
be able to identify flow asynchrony from waveforms
Scoop waveform
VC/AC mode
Ventilator mode where flow asynchrony is commonly seen
Increase the inspiratory flow
Decrease in I time and increase in E time
Decrease the inspiratory flow
Increase in I time and decrease in E time
For adults, peak flow values are generally kept between 60-80 L/min
For adults, the typical I-time is 0.8 to 1 second
Square waveform
Lower peak flow and decreased inspiratory time, compared to decelerating
Decelerating waveform
Mimics more natural breathing pattern, produces high initial peak flow but increased iT, pts are usually placed on this waveform on VC AC mode
Rise time
Allows adjustment of the rate
Rapid rise time
Peak is created showing flow being delivered too fast and too high
Appropriate rise time
Just right
Slow rise time
Curve is too low
Modes that utilize rise time
PC & VC+AC
PC & VC+ SIMV
PSV
VS
Cycle asynchrony
Occurs when the ventilator does not match the patient breathing pattern
Long cycling
iT is too long, excessive and pt may attempt to actively exhale while the ventilator is still in the inspiratory phase resulting in a spike in airway pressure
Short (double) cycling
iT is too short due to premature ventilator cycling
The parameter that allows the patient to exhale while they are on PSV mode is 25%
Mode asynchrony
Occurs when the mode of ventilation selected is unable to match pts spontaneous vent pattern
Patient-Ventilator Synchrony
Adequate ventilation and oxygenation, decrease WOB and patient comfort
Patient-Ventilator Dysynchrony
Ineffective oxygenation and ventilation, increase WOB and patient discomfort
Increase in ventilators workload
Can lead to ventilatory muscle fatigue and structural injury of the lung
It takes 24 hours to recover from ventilatory muscle fatigue