MECH VENT 2

    Subdecks (12)

    Cards (333)

    • Flow asynchrony
      Inspiratory gas flow from the vent is unable to match the patients inspiratory flow demand
    • be able to identify flow asynchrony from waveforms
      • Scoop waveform
    • VC/AC mode
      Ventilator mode where flow asynchrony is commonly seen
    • Increase the inspiratory flow

      Decrease in I time and increase in E time
    • Decrease the inspiratory flow

      Increase in I time and decrease in E time
    • For adults, peak flow values are generally kept between 60-80 L/min
    • For adults, the typical I-time is 0.8 to 1 second
    • Square waveform
      Lower peak flow and decreased inspiratory time, compared to decelerating
    • Decelerating waveform
      Mimics more natural breathing pattern, produces high initial peak flow but increased iT, pts are usually placed on this waveform on VC AC mode
    • Rise time
      Allows adjustment of the rate
    • Rapid rise time

      • Peak is created showing flow being delivered too fast and too high
    • Appropriate rise time
      • Just right
    • Slow rise time
      • Curve is too low
    • Modes that utilize rise time
      • PC & VC+AC
      • PC & VC+ SIMV
      • PSV
      • VS
    • Cycle asynchrony
      Occurs when the ventilator does not match the patient breathing pattern
    • Long cycling
      iT is too long, excessive and pt may attempt to actively exhale while the ventilator is still in the inspiratory phase resulting in a spike in airway pressure
    • Short (double) cycling
      iT is too short due to premature ventilator cycling
    • The parameter that allows the patient to exhale while they are on PSV mode is 25%
    • Mode asynchrony
      Occurs when the mode of ventilation selected is unable to match pts spontaneous vent pattern
    • Patient-Ventilator Synchrony
      Adequate ventilation and oxygenation, decrease WOB and patient comfort
    • Patient-Ventilator Dysynchrony
      Ineffective oxygenation and ventilation, increase WOB and patient discomfort
    • Increase in ventilators workload

      Can lead to ventilatory muscle fatigue and structural injury of the lung
    • It takes 24 hours to recover from ventilatory muscle fatigue
    • Dangers of Controlled Ventilation
      Ventilatory muscle weakness, deconditioning, atrophy
    • Things that can worsen patient-ventilator asynchrony
      • Pain
      • Anxiety
      • Procedural interactions
      • Other clinician-patient interactions
    • Types of patient-ventilator asynchrony
      • Trigger asynchrony
      • Flow asynchrony
      • Cycle asynchrony
      • Mode asynchrony
    • Trigger Asynchrony
      Vent is not triggering breath to patient even though pt demands it
    • Flow Asynchrony
      Poor coordination between the pts flow demand and the flow provided by the vent. Common in VC.
    • Cycle Asynchrony
      Poor coordination between ventilator and pt respiratory drive
    • Mode Asynchrony
      Asynchrony develops when the mode selected does not match the pts spontaneous vent efforts
    • Trigger
      Parameter on the ventilator that initiates a breath
    • Types of trigger
      • Time trigger
      • Flow trigger
      • Pressure trigger
    • Patient-triggered
      Pressure trigger or flow trigger
    • Controlled Breaths
      Time trigger
    • There would be no dips or jumps on none of the waveforms to identify a time trigger breath
    • When will a patient receive a time trigger breath
      Once the preset time interval is met. 60sec/RR=preset time interval
    • A jump at the end of exhalation will be shown on the flow-time scalar to identify a flow trigger breath
    • How the ventilator knows to deliver a flow trigger breath
      When a change in bias flow is detected
    • Flow trigger
      Usually set between 1 to 2 L/min
    • There is a dip at the beginning of the pressure-time scalar waveform to identify a pressure trigger breath