MECH VENT 2

Subdecks (12)

Cards (333)

  • Flow asynchrony
    Inspiratory gas flow from the vent is unable to match the patients inspiratory flow demand
  • be able to identify flow asynchrony from waveforms
    • Scoop waveform
  • VC/AC mode
    Ventilator mode where flow asynchrony is commonly seen
  • Increase the inspiratory flow

    Decrease in I time and increase in E time
  • Decrease the inspiratory flow

    Increase in I time and decrease in E time
  • For adults, peak flow values are generally kept between 60-80 L/min
  • For adults, the typical I-time is 0.8 to 1 second
  • Square waveform
    Lower peak flow and decreased inspiratory time, compared to decelerating
  • Decelerating waveform
    Mimics more natural breathing pattern, produces high initial peak flow but increased iT, pts are usually placed on this waveform on VC AC mode
  • Rise time
    Allows adjustment of the rate
  • Rapid rise time

    • Peak is created showing flow being delivered too fast and too high
  • Appropriate rise time
    • Just right
  • Slow rise time
    • Curve is too low
  • Modes that utilize rise time
    • PC & VC+AC
    • PC & VC+ SIMV
    • PSV
    • VS
  • Cycle asynchrony
    Occurs when the ventilator does not match the patient breathing pattern
  • Long cycling
    iT is too long, excessive and pt may attempt to actively exhale while the ventilator is still in the inspiratory phase resulting in a spike in airway pressure
  • Short (double) cycling
    iT is too short due to premature ventilator cycling
  • The parameter that allows the patient to exhale while they are on PSV mode is 25%
  • Mode asynchrony
    Occurs when the mode of ventilation selected is unable to match pts spontaneous vent pattern
  • Patient-Ventilator Synchrony
    Adequate ventilation and oxygenation, decrease WOB and patient comfort
  • Patient-Ventilator Dysynchrony
    Ineffective oxygenation and ventilation, increase WOB and patient discomfort
  • Increase in ventilators workload

    Can lead to ventilatory muscle fatigue and structural injury of the lung
  • It takes 24 hours to recover from ventilatory muscle fatigue
  • Dangers of Controlled Ventilation
    Ventilatory muscle weakness, deconditioning, atrophy
  • Things that can worsen patient-ventilator asynchrony
    • Pain
    • Anxiety
    • Procedural interactions
    • Other clinician-patient interactions
  • Types of patient-ventilator asynchrony
    • Trigger asynchrony
    • Flow asynchrony
    • Cycle asynchrony
    • Mode asynchrony
  • Trigger Asynchrony
    Vent is not triggering breath to patient even though pt demands it
  • Flow Asynchrony
    Poor coordination between the pts flow demand and the flow provided by the vent. Common in VC.
  • Cycle Asynchrony
    Poor coordination between ventilator and pt respiratory drive
  • Mode Asynchrony
    Asynchrony develops when the mode selected does not match the pts spontaneous vent efforts
  • Trigger
    Parameter on the ventilator that initiates a breath
  • Types of trigger
    • Time trigger
    • Flow trigger
    • Pressure trigger
  • Patient-triggered
    Pressure trigger or flow trigger
  • Controlled Breaths
    Time trigger
  • There would be no dips or jumps on none of the waveforms to identify a time trigger breath
  • When will a patient receive a time trigger breath
    Once the preset time interval is met. 60sec/RR=preset time interval
  • A jump at the end of exhalation will be shown on the flow-time scalar to identify a flow trigger breath
  • How the ventilator knows to deliver a flow trigger breath
    When a change in bias flow is detected
  • Flow trigger
    Usually set between 1 to 2 L/min
  • There is a dip at the beginning of the pressure-time scalar waveform to identify a pressure trigger breath