One of two photographic properties comprising detail visibility, the other being brightness/density
Synonyms of contrast
Dynamic Range
Window Width
Radiographic Contrast
The difference between two or more brightness levels or densities on a radiograph, resulting from tissue's differential absorption of x-ray photons
Lack of contrast makes an image appear homogenous
Radiographic contrast
Affects visibility of structural lines making up recorded images
Results from multiple factors: anatomic structure, quality of radiation, image receptor type, computer processing and display
Subject Contrast
A result of the absorption characteristics of the anatomic tissue radiographed along with the quality of the x-ray beam
Types of contrast
Low Contrast (Few shades of gray, Increased contrast, Low kVp, Short scale)
High Contrast (Many shades of gray, Decreased contrast, High kVp, Long scale)
Long scale
Decreased contrast, Higher kV used, Many shades of gray, Gradual changes in tones
Short scale
Increased contrast, Lower kV used, Fewer shades of gray, Black & white
This chest x-ray appears to be a long scale: many shades of gray, gradual changes in tones, higher kV used
Radiographic contrast
Results from subject contrast and IR contrast
Subject Contrast
The ratio of the x-ray intensity transmitted through one segment of the part in a study to that transmitted through another part, affected by part thickness, atomic number, radiation quality, scatter radiation, use of contrast agents
Contrast Resolution
The ability of the system to distinguish between small objects that attenuate the x-ray beam similarly, specifically used in digital imaging
The human eye has limited contrast resolution
Window width
A computer processing tool that adjusts radiographic contrast
Grayscale
The number of different shades of gray that can be stored and displayed by a computer system, determined by bit depth
Wide window width
Lower contrast
Narrow window width
Higher contrast
Kilovoltage Peak (kVp)
The penetrating power of the x-ray beam, alters the quantity and quality of the x-ray beam, changes radiation exposure to the IR, important for image quality in digital radiography
Increase in kVp
Increase in radiation exiting the patient, increase in scatter radiation, decrease in contrast
Decrease in kVp
Decrease in radiation exiting the patient, decrease in scatter radiation, increase in contrast
In digital radiography, the computer will post process the image making any over- or under-usage of kVp mostly unnoticeable
Too much radiation
Image appears appropriately bright, but patient is overexposed
Too little radiation
Image appears appropriately bright, but quantum noise is increased which decreases detail
When altering mAs is limited, the kVp must be changed to maintain exposure to the IR
The 15% Rule - changing the kVp by 15% has the same effect as doubling or halving the mAs
Increasing kVp
Decreased contrast
Decreasing kVp
Increased contrast
Radiographs not generally repeated due to contrast errors, if a repeat is necessary to change the contrast: adjust the kVp by 15%, and adjust mAs by an inverse factor of 2
A higher kVp and a lower mAs should be used to reduce patient dose
Secondary factors affecting radiographic contrast
OID
Collimation
Filtration
Grid
Anatomic part
Focal spot size
Anode heel effect
OID (Object-to-Image Distance)
The distance between the object being radiographed and the image receptor, increasing OID reduces scatter radiation reaching the IR, increasing contrast
Collimation
Decreasing the field of view, reduces x-ray photon quantity and scatter, increasing contrast
Filtration
Filters low-energy x-ray photons, leaving the x-ray beam to consist of mostly high-energy photons, high-energy = increased scatter = decreased contrast
Filtration
Inversely related to contrast, as filtration increases, contrast decreases
Grids
Absorb scatter radiation from the remnant beam, increase in grid ratio = decrease in scatter reaching IR = increase in contrast, but using a grid does increase patient dose
Anatomic part size
Inversely related to contrast, as part size increases, scatter produced increases, causing contrast to decrease
Tissue density/atomic number
Directly related to contrast, as the atomic number of tissue increases, more photoelectric absorption occurs, causing contrast to increase
Focal spot size changes spatial resolution but does not affect contrast, and the Anode Heel Effect has little to no effect on contrast
two photographic properties comprising detail visibility: