Steps in calibrating using the micrometers
1. The eyepiece is first removed from the draw tube
2. The lens cover of the eyepiece is then unscrewed
3. The ocular micrometer is inserted
4. The lens cover is replaced and the eyepiece is returned to the draw tube
5. For checking, look into the eyepiece to check whether the disc is not inverted
6. The stage micrometer is placed on the stage and the scale is focused
7. The stage micrometer is focused under specific objective that will be utilized to measure the specimen size
8. The SM is arranged so that a line on the far left should correspond to a line in the OM. It should be overlapping
9. After coinciding the first line, another line to the right should coincide with that of another line in the OM