quality assurance

Cards (31)

  • Quality Assurance and Quality Control

    • Managing Patient Care and Image Quality
  • Quality Control
    • Equipment quality control measures
    • Acceptance testing
    • Performance monitoring
    • Maintenance (including preventive maintenance)
  • QC measures are also tied to accreditation and reimbursement
  • Most states have legislation that requires QC measures to be performed and documented
  • Radiographic Quality Control
    • Filtration
    • Collimated field/x-ray field congruence
    • SID indicator
    • Light beam centering device
    • Focal-spot size
    • kVp accuracy
    • Timer accuracy
    • Linearity
    • Reproducibility
  • Inherent + added filtration
    >2.5 mm Al for tubes operated over 70 kVp
  • Collimated field/x-ray field congruence
    +/- 2% of the SID
  • SID indicator
    +/- 2% of the SID
  • Light beam centering device

    +/- 1% of SID
  • Focal-spot size

    <150% of nominal size
  • kVp accuracy
    +/- 10% of indicated kVp
  • Timer accuracy
    +/- 5% over 10 msec
    +/- 20% <10 msec
  • Linearity
    +/- 10% for adjacent mA stations
  • Reproducibility
    +/- 5% for sequential exposures
  • The machine does not meet reproducibility requirements
  • The machine passes the linearity test
  • Aprons, gloves and shields are examined under fluoroscopy or radiographed annually for cracks, holes or defects
  • Defective lead apparel is replaced
  • Digital Imaging
    • Detector/CR cassette performance
    • Software performance
    • Monitor display
    • DICOM and GSDF
  • Fluoroscopic Quality Control

    • Exposure rate (ESE) not to exceed 10 R/min
    High dose fluoroscopy for interventional or cardiac may not exceed 20 R/min
    ABC systems (AERC) must be calibrated for II and FPD
  • Mammography as a Model for QC
    • Annual QC evaluations
    Elements of a mammographic QC program
  • Exposure Technique Charts
    Purpose is to allow for consistent quality image production
    Decreases repeat rate
    Lowers patient dose
  • Manufacturers often provide technique charts with equipment
  • These charts are guidelines and should not be used without modification
  • Technique charts may vary between different radiographic rooms in the same department
  • Types of Technique Charts
    • Variable kVp chart
    Fixed kVp chart
    High kVp chart
    AEC chart
  • Variable kVp Technique Chart
    Uses a fixed mAs and varies the kVp according to the thickness of the part
    Advantage: the use of a lower kVp results in a shorter scale of contrast
    Disadvantages: may result in higher patient dose and less exposure latitude
  • Variable kVp Technique Chart formula

    Beginning kVp (high frequency generator) = 2 x thickness of part (cm) + 23
    For single phase add 30 and for three phase add 25
    The kVp increases by 2 kVp/cm
    Make test exposures of a phantom with three different mAs values and pick the one with optimum exposure
  • Fixed kVp Technique Chart
    Select an optimum kVp for penetration of part
    Vary mAs based on size of part/patient
    Patients are sorted into small, medium and large categories
    Advantage: may allow for lower patient dose and more exposure latitude due to inherently higher kVp
    mAs changes must be between 30-50%
  • High kVp Technique Chart
    Used for barium studies and chest radiography
    kVp is over 100
    Lower patient dose and increased exposure latitude
  • AEC Technique Chart
    Positioning is critical
    May be APR system of variable kVp system, or both
    kVp must be specified or programmed
    Density control buttons specified
    Collimation specified
    Detector cells specified