quality assurance

    Cards (31)

    • Quality Assurance and Quality Control

      • Managing Patient Care and Image Quality
    • Quality Control
      • Equipment quality control measures
      • Acceptance testing
      • Performance monitoring
      • Maintenance (including preventive maintenance)
    • QC measures are also tied to accreditation and reimbursement
    • Most states have legislation that requires QC measures to be performed and documented
    • Radiographic Quality Control
      • Filtration
      • Collimated field/x-ray field congruence
      • SID indicator
      • Light beam centering device
      • Focal-spot size
      • kVp accuracy
      • Timer accuracy
      • Linearity
      • Reproducibility
    • Inherent + added filtration
      >2.5 mm Al for tubes operated over 70 kVp
    • Collimated field/x-ray field congruence
      +/- 2% of the SID
    • SID indicator
      +/- 2% of the SID
    • Light beam centering device

      +/- 1% of SID
    • Focal-spot size

      <150% of nominal size
    • kVp accuracy
      +/- 10% of indicated kVp
    • Timer accuracy
      +/- 5% over 10 msec
      +/- 20% <10 msec
    • Linearity
      +/- 10% for adjacent mA stations
    • Reproducibility
      +/- 5% for sequential exposures
    • The machine does not meet reproducibility requirements
    • The machine passes the linearity test
    • Aprons, gloves and shields are examined under fluoroscopy or radiographed annually for cracks, holes or defects
    • Defective lead apparel is replaced
    • Digital Imaging
      • Detector/CR cassette performance
      • Software performance
      • Monitor display
      • DICOM and GSDF
    • Fluoroscopic Quality Control

      • Exposure rate (ESE) not to exceed 10 R/min
      High dose fluoroscopy for interventional or cardiac may not exceed 20 R/min
      ABC systems (AERC) must be calibrated for II and FPD
    • Mammography as a Model for QC
      • Annual QC evaluations
      Elements of a mammographic QC program
    • Exposure Technique Charts
      Purpose is to allow for consistent quality image production
      Decreases repeat rate
      Lowers patient dose
    • Manufacturers often provide technique charts with equipment
    • These charts are guidelines and should not be used without modification
    • Technique charts may vary between different radiographic rooms in the same department
    • Types of Technique Charts
      • Variable kVp chart
      Fixed kVp chart
      High kVp chart
      AEC chart
    • Variable kVp Technique Chart
      Uses a fixed mAs and varies the kVp according to the thickness of the part
      Advantage: the use of a lower kVp results in a shorter scale of contrast
      Disadvantages: may result in higher patient dose and less exposure latitude
    • Variable kVp Technique Chart formula

      Beginning kVp (high frequency generator) = 2 x thickness of part (cm) + 23
      For single phase add 30 and for three phase add 25
      The kVp increases by 2 kVp/cm
      Make test exposures of a phantom with three different mAs values and pick the one with optimum exposure
    • Fixed kVp Technique Chart
      Select an optimum kVp for penetration of part
      Vary mAs based on size of part/patient
      Patients are sorted into small, medium and large categories
      Advantage: may allow for lower patient dose and more exposure latitude due to inherently higher kVp
      mAs changes must be between 30-50%
    • High kVp Technique Chart
      Used for barium studies and chest radiography
      kVp is over 100
      Lower patient dose and increased exposure latitude
    • AEC Technique Chart
      Positioning is critical
      May be APR system of variable kVp system, or both
      kVp must be specified or programmed
      Density control buttons specified
      Collimation specified
      Detector cells specified