What is the difference between unipolar and bipolar stimulation?
Location of the anode
Unipolar = the anode is the case of the pulse generator
Bipolar = the anode is located on the pacing lead (2cm proximal)
The cathode in both systems is at the lead tip in contact with the myocardium
Why is the impedance and threshold slightly higher in bipolar systems?
There are 2 conductors in one lead, compared to the single conductor unipolar lead
How can you identify the difference between bipolar and unipolar pacing on an ECG?
Pacing spikes!
Unipolar = BIG pacing spikes, useful for identifying pacing (mal)function
Bipolar = small pacing spikes, often difficult to see on ECG
How does a unipolar circuit work?
Impulse begins at generator (anode)
Flows through the lead tip (cathode)
Stimulates heart
Returns through tissue and body fluid to anode
What are the characteristics of a unipolar pacing lead?
One electrode at tip
One conductor coil
Smaller in diameter than bipolar
Lots of artefact!
How does a bipolar circuit work?
Impulse begins at generator
Flows through lead tip (cathode)
Stimulates heart
Returns to ring electrode above lead tip (anode)
What are the characteristics of a bipolar pacing lead?
Creates a circuit between ring and tip electrode
Two conductor coils
Inner layer of insulation
Less susceptible to oversensing of non-cardiac signals
What type of lead is used as standard practice?
Bipolar
Can still be reprogrammed in unipolar configuration when required but often has fewer issues than a genuine unipolar circuit
What is an intracardiac electrogram (iECG)?
Electrical signal from within the heart, produced by movement of electrical current through the myocardium
Pacing inhibition and shock delivery rely on iECG!
Able to distinguish between SVT and VT via presence or absence of relationship between atrium and ventricles
Also identifies FFRWS and skeletal muscle signals
How does depolarisation relate to iECGs?
In depolarisation, the outside of the cell is electrically neutral with respect to inside
As wavefront of depolarisation travels TOWARDS endocardial electrode, it becomes positively charged
This manifests as a POSITIVE deflection
As wavefront of depolarisation travels UNDER electrode, cell becomes negatively charged
This manifests as a NEGATIVE deflection
The intrinsic deflection indicates the moment of activation under recording electrode
Why is the ventricular electrogram bigger than the atrial?
Greater mass of myocardium
Hence more prone to FFRW sensing
What is Fourier transformation?
Frequency spectrum expressed as a series of sine waves of varying frequency and amplitude
Maximum density for R waves = 10-30Hz
Filtering below 10Hz (low frequency) reduces force of T wave/FFRWS/myopotentials
For a signal to be sensed, it must be of sufficient amplitude and the intrinsic deflection must have sufficient slope
What is the typical frequency for iECGs?
30Hz (up to 40Hz)
How does bipolar and unipolar sensing work?
Measures the difference between 2 electrodes
Bipolar = interelectrode difference of 2-3cm - both electrode in heart
Unipolar = interelectrode difference of 30-50cm - one electrode in heart, one in pulse generator
Both electrodes contribute to the sensed signal... lots of influence from non-cardiac signals
What are the benefits of bipolar sensing?
Mostly immune to myopotentials... unless break in circuit
Less influenced by EMI (e.g. diathermy or cautery)
More sensitive to direction of depolarisation waveform, both electrodes are activated at the same time when wave is perpendicular to iECG (and so is often biphasic)
Why can you not use unipolar leads in ICDs?
ICDs are more sensitive and have short refractory periods to interrupt tachyarrhythmias and reduce risk of TWOS
Oversensing can lead to inappropriate shocks
Must use bipolar leads!
What is an afterdepolarisation?
Following a paced beat, an after potential of opposite charge is induced in myocardium at the electrode tip
Positive ions surround electrode after a (negative) cathodal stimulation
The after potential can be inappropriately sensed = inhibition of next paced beat
Amplitude of after potential directly proportional to amplitude and duration of paced beat
Much more likely to occur at high output pacing and/or a long pulse width!
What refractory period prevents the inappropriate sensing of afterpotentials?
PVAB - prevents inhibition and/or triggering. Signal is blanked
What is a dangerous complication of inappropriately sensing afterpotentials?
Crosstalk
Inhibition of pacing
Signal may be sensed in the opposite chamber in DDD systems
What is an injury current?
ST segment elevation on iECG due to pressure exerted by distal electrode on myocardium
Occurs in both atrial and ventricular leads, in both active and passive fixation
The absence of an injury current usually indicates poor electrode contact with myocardium
Area of fibrotic or scarred myocardium?
Returns to isoelectric line shortly after (hours-days post implant)
Often much more intense in active fixation due to screw-in mechanism
What are the 5 components of lead design?
Electrode(s)
Conductors (s)
Insulation
Connector pin
Fixation mechanism
Electrodes
The smaller the radius of the electrode, the greater the current density
Resistance is higher with smaller electrodes
Enables constant voltage pacing with improved battery longevity
Sensing impedance and afterdepolarisation are decreased with larger surface area
The ideal pacing lead would have a SMALL radius (increased current density) and a LARGE surface area (decreased polarisation)
Solution = shape!
What are the features of modern pacing leads?
Small electrode tip with a complex surface structure
Screw shaped, helical, hooks and barbs used
Complex shapes produce an irregular pattern of current density at edges of lead tip, whilst covering a large surface area
What is the surface structure of an early vs. modern pacing lead?
Early = polished metal surface
Modern = textured surface - allowing an increase in surface area without an increase in radius
How can lead design reduce polarisation?
Pores on lead tip
Large (130 microns)
Microscopic (20-100 microns)
Coating
Elgiroy / platinum / iridium oxide
Carbon electrodes
Roughening of surface of lead tip, known as activation
Sensing improved by porous electrode design, but threshold generally improved by active fixation and steroid eluting tips
What is the chemical composition used in new pacing leads?
Platinum-iridium / elgiroy / titanium coated with platinum or carbon (improved with activation)
Electrodes must be biologically inert and resistant to degradation, to minimise inflammation and fibrosis
Metals such as zinc, copper, mercury, lead, nickel and silver = toxic
Stainless steel = corrosion
Titanium requires a surface coating of oxides, which may impede charge transfer
However, coated with carbon or platinum = excellent long term!
What electrodes do each manufacturer use?
Medtronic = platinum iridium with polyurethane outer jacket
Boston = iridium oxide
Abbott = titanium nitrade
What steroid is used in steroid eluting electrodes?
Dexamethasone sodium phosphate
Silicone core impregnated with dexamethasone, surrounded by a porous titanium electrode and coated with platinum
What are the benefits of using steroid eluting electrodes?
Reduces inflammation at myocardial surface
Controls chronic evolution of thresholds over time
Reduces risk of exit block
Reduces risk of calcification or fibrosis
Duration of drug elution not defined, but evidence of reduced myocardial inflammation for 10+ years!
What are the characteristics of passive leads?
Tines
Trapped in trabeculae immediately upon correct positioning
Rapidly covered in fibrous tissue, very difficult to remove after 6 months - more difficult than active!