  • What is the difference between unipolar and bipolar stimulation?
    Location of the anode
    • Unipolar = the anode is the case of the pulse generator
    • Bipolar = the anode is located on the pacing lead (2cm proximal)
    The cathode in both systems is at the lead tip in contact with the myocardium
  • Why is the impedance and threshold slightly higher in bipolar systems?
    There are 2 conductors in one lead, compared to the single conductor unipolar lead
  • How can you identify the difference between bipolar and unipolar pacing on an ECG?
    Pacing spikes!
    • Unipolar = BIG pacing spikes, useful for identifying pacing (mal)function
    • Bipolar = small pacing spikes, often difficult to see on ECG
  • How does a unipolar circuit work?
    1. Impulse begins at generator (anode)
    2. Flows through the lead tip (cathode)
    3. Stimulates heart
    4. Returns through tissue and body fluid to anode
  • What are the characteristics of a unipolar pacing lead?
    1. One electrode at tip
    2. One conductor coil
    3. Smaller in diameter than bipolar
    4. Lots of artefact!
  • How does a bipolar circuit work?
    1. Impulse begins at generator
    2. Flows through lead tip (cathode)
    3. Stimulates heart
    4. Returns to ring electrode above lead tip (anode)
  • What are the characteristics of a bipolar pacing lead?
    1. Creates a circuit between ring and tip electrode
    2. Two conductor coils
    3. Inner layer of insulation
    4. Less susceptible to oversensing of non-cardiac signals
  • What type of lead is used as standard practice?
    Bipolar
    • Can still be reprogrammed in unipolar configuration when required but often has fewer issues than a genuine unipolar circuit
  • What is an intracardiac electrogram (iECG)?
    Electrical signal from within the heart, produced by movement of electrical current through the myocardium
    • Pacing inhibition and shock delivery rely on iECG!
    • Able to distinguish between SVT and VT via presence or absence of relationship between atrium and ventricles
    • Also identifies FFRWS and skeletal muscle signals
  • How does depolarisation relate to iECGs?
    • In depolarisation, the outside of the cell is electrically neutral with respect to inside
    • As wavefront of depolarisation travels TOWARDS endocardial electrode, it becomes positively charged
    • This manifests as a POSITIVE deflection
    • As wavefront of depolarisation travels UNDER electrode, cell becomes negatively charged
    • This manifests as a NEGATIVE deflection
    • The intrinsic deflection indicates the moment of activation under recording electrode
  • Why is the ventricular electrogram bigger than the atrial?
    Greater mass of myocardium
    • Hence more prone to FFRW sensing
  • What is Fourier transformation?

    Frequency spectrum expressed as a series of sine waves of varying frequency and amplitude
    • Maximum density for R waves = 10-30Hz
    • Filtering below 10Hz (low frequency) reduces force of T wave/FFRWS/myopotentials
    • For a signal to be sensed, it must be of sufficient amplitude and the intrinsic deflection must have sufficient slope
  • What is the typical frequency for iECGs?
    30Hz (up to 40Hz)
  • How does bipolar and unipolar sensing work?
    Measures the difference between 2 electrodes
    • Bipolar = interelectrode difference of 2-3cm - both electrode in heart
    • Unipolar = interelectrode difference of 30-50cm - one electrode in heart, one in pulse generator
    Both electrodes contribute to the sensed signal... lots of influence from non-cardiac signals
  • What are the benefits of bipolar sensing?
    • Mostly immune to myopotentials... unless break in circuit
    • Less influenced by EMI (e.g. diathermy or cautery)
    • More sensitive to direction of depolarisation waveform, both electrodes are activated at the same time when wave is perpendicular to iECG (and so is often biphasic)
  • Why can you not use unipolar leads in ICDs?
    ICDs are more sensitive and have short refractory periods to interrupt tachyarrhythmias and reduce risk of TWOS
    • Oversensing can lead to inappropriate shocks
    Must use bipolar leads!
  • What is an afterdepolarisation?
    Following a paced beat, an after potential of opposite charge is induced in myocardium at the electrode tip
    • Positive ions surround electrode after a (negative) cathodal stimulation
    • The after potential can be inappropriately sensed = inhibition of next paced beat
    • Amplitude of after potential directly proportional to amplitude and duration of paced beat
    • Much more likely to occur at high output pacing and/or a long pulse width!
  • What refractory period prevents the inappropriate sensing of afterpotentials?
    PVAB - prevents inhibition and/or triggering. Signal is blanked
  • What is a dangerous complication of inappropriately sensing afterpotentials?
    Crosstalk
    • Inhibition of pacing
    • Signal may be sensed in the opposite chamber in DDD systems
  • What is an injury current?
    ST segment elevation on iECG due to pressure exerted by distal electrode on myocardium
    • Occurs in both atrial and ventricular leads, in both active and passive fixation
    • The absence of an injury current usually indicates poor electrode contact with myocardium
    • Area of fibrotic or scarred myocardium?
    • Returns to isoelectric line shortly after (hours-days post implant)
    • Often much more intense in active fixation due to screw-in mechanism
  • What are the 5 components of lead design?
    1. Electrode(s)
    2. Conductors (s)
    3. Insulation
    4. Connector pin
    5. Fixation mechanism
  • Electrodes
    The smaller the radius of the electrode, the greater the current density
    • Resistance is higher with smaller electrodes
    • Enables constant voltage pacing with improved battery longevity
    • Sensing impedance and afterdepolarisation are decreased with larger surface area
    • The ideal pacing lead would have a SMALL radius (increased current density) and a LARGE surface area (decreased polarisation)
    • Solution = shape!
  • What are the features of modern pacing leads?

    Small electrode tip with a complex surface structure
    • Screw shaped, helical, hooks and barbs used
    • Complex shapes produce an irregular pattern of current density at edges of lead tip, whilst covering a large surface area
  • What is the surface structure of an early vs. modern pacing lead?

    Early = polished metal surface
    Modern = textured surface - allowing an increase in surface area without an increase in radius
  • How can lead design reduce polarisation?
    • Pores on lead tip
    • Large (130 microns)
    • Microscopic (20-100 microns)
    • Coating
    • Elgiroy / platinum / iridium oxide
    • Carbon electrodes
    • Roughening of surface of lead tip, known as activation
  • Sensing improved by porous electrode design, but threshold generally improved by active fixation and steroid eluting tips
  • What is the chemical composition used in new pacing leads?

    Platinum-iridium / elgiroy / titanium coated with platinum or carbon (improved with activation)
    • Electrodes must be biologically inert and resistant to degradation, to minimise inflammation and fibrosis
    • Metals such as zinc, copper, mercury, lead, nickel and silver = toxic
    • Stainless steel = corrosion
    • Titanium requires a surface coating of oxides, which may impede charge transfer
    • However, coated with carbon or platinum = excellent long term!
  • What electrodes do each manufacturer use?
    Medtronic = platinum iridium with polyurethane outer jacket
    Boston = iridium oxide
    Abbott = titanium nitrade
  • What steroid is used in steroid eluting electrodes?
    Dexamethasone sodium phosphate

    Silicone core impregnated with dexamethasone, surrounded by a porous titanium electrode and coated with platinum
  • What are the benefits of using steroid eluting electrodes?
    Reduces inflammation at myocardial surface
    • Controls chronic evolution of thresholds over time
    • Reduces risk of exit block
    • Reduces risk of calcification or fibrosis
    • Duration of drug elution not defined, but evidence of reduced myocardial inflammation for 10+ years!
  • What are the characteristics of passive leads?
    • Tines
    • Trapped in trabeculae immediately upon correct positioning
    • Rapidly covered in fibrous tissue, very difficult to remove after 6 months - more difficult than active!
    • Minimises trauma to myocardium... smaller fibrous capsule = lower chronic threshold
    • Larger introducer (7F+) required due to increased diameter of lead
  • What are the characteristics of active leads?
    • Screw helix
    • Most common = extendable-retractable design
    • Allows screw helix to extend from tip once through the vascular system and over tricuspid valve = less trauma
    • Especially useful in atrium, provides stable fixation even in AF or post CABG when atrial appendage is injured
    • Test in deployed and non-deployed at implant
  • Why was the J-retention wire discontinued?
    Very easy to implant, but...
    • Caused metal fatigue from flexing duration cardiac motion
    • Frequent fractures
    • Increased risk of tamponade and perforation
  • What is a lead conductor (DBS/drawn brazed strand)?

    Wire between electrode tip and connector pin
    • At 70bpm, heart contracts 36 million times per year = significant stress on lead!
    • Leads must flex with heartbeat and have longitudinal/transverse/rotational components
    • Fracture and metal fatigue can occur
    • Bipolar leads = COAXIAL design
  • What is a coaxial design?

    Occurs in bipolar leads
    • Conductor coil to distal electrode is INSIDE the outer conductor coil
    • Inner coil connects to tip electrode
    • Outer coil connects to ring electrode
    • Sleeve of insulation separates the two coils
    • Can make lead less flexible than unipolar leads
    • Remember, unipolar leads only have a single conductor coil!
  • What are the characteristics of silicone insulated leads?
    • Very flexible
    • Repairable
    • Easy to mould into shape
    • Tears and cuts easily
    • High friction in blood (this was an old exam question!!)
    • Absorbs lipids - prone to calcification and fibrosis
    • Larger diameter, requires bigger sheath
    • 30+ year history of success
  • What are the characteristics of polyurethane insulated leads?
    • Stiff
    • Not repairable
    • Low friction in blood
    • High resistance to tears and cuts
    • Small diameter, smaller sheath required
    • Sensitive manufacturing process - cooling
    • Prone to cracking from environmental stress - suturing, stylet insertion, oxidisation
    • 10+ year history of success
  • What insulation do the manufacturers use?

    Medtronic = polyurethane
    Boston = silicone
    Abbott = 'optim' design - hybrid of silicone and polyurethane
  • What is the diameter of a standard IS-1 lead?
    3.2mm with no sealing rings in header
  • Why were connector pins incompatible?

    No clinical standard
    • Mix of lead connectors and pulse generator headers
    • Evolved from 5.6mm to 3.2mm (IS-1)
    • Some old leads may not fit in all headers
    • E.g. old Medtronic leads require IS-1B with a long receptacle for lead terminal.