Precursor Ion ---> Transition (Q1) ---> MS Detector
Selected Ion Monitoring:
Precursor Ion ---> Filter for Precursor Ion (Q1) ---> Transition (Q2) --- MS Detector
Multiple Reaction Monitoring:
Precursor Ion ---> First Filter for Precursor Ion (Q1) ---> Fragmentation in Collision Cell (Using Argon or Helium)(Q2) ---> Second Filter for Product Ion (Q3) ---> MS Detector
Parallel Reaction Monitoring:
Precursor Ion ---> Filter for Precursor Ion ---> Fragmentation in the CollisionCell (Using Argon or Helium) (Q2) ---> Orbitrap (MS Detector)
Characteristics:
Ion Transmission - for sensitivity
Scan Speed - Number of MSexperiments that can be done
Mass Accuracy - difference between the theoretical and measured mass
Mass Resolution - separation of narrow mass spectral peaks
Resolving Power - degree to whichseparation of narrow peaks can occur
Mass Range Limit - Low Resolution can do up to 2000 m/z but High Resolution can do up to 200,000
Resolving Power (Full Width at Half Maximum) Ranking
At the top:
FT-ICR-MS - 1,000,000
FT-Orbitrap
High-Res-TOF - 6000
TOF
Quadrupole and Ion Trap (ultrazoom mode)
Quadrupole /Ion Trap - 1000
Ion Transmission:
How better the transfer of ions from the source to the detector through the mass analyser.
ion optics
ion guides
hexapoles or octupoles
t-wave
step-wave
Quadrupole Ion Trap 3D:
doughnut-shaped to confine space
does a pre-scan to estimate number of ions
only space for so many
ions will repel
Quadrupole Ion Trap Linear:
Box-shaped
better for quantitation than 3D
less chancing of gating
Things to Consider when selecting Operation Mode:
sensitivity
Reduction of noise
Selectivity
Speed
Selected Ion Monitoring:
ions of interest monitored
less discriminating
high resolution
better S:N (3:1)
Multiple Reaction Monitoring:
ions of interest
longer focus on precursor and fragment ions (good sensitivity and selectivity)
Duty Cycle - proportion of time ions are accepted from an ion source and transferredthrough to the detection (time from ion source to detector)
Scan Speed - time taken for RF (AC and/or DC) potentials to be applied
Dwell Time - time taken for two quadrupoles (Q1 and Q2 or Q1 and Q3) to scan through the transition
Tandem in Time:
Quadrupole Ion Trap
Linear Ion Trap
IonCyclotronResonance
Orbitrap
Tandem in Space:
Triple Quadrupole
High Resolution MagneticSector
TOF/TOF
Types of Ion Dissociation:
Metastable Ions - spontaneously dissociate en route from the ion source to the detector
Collisionally Activated Dissociation - ions activated by collision with neutral gases (neutral losses and product ion)
Electron Capture Dissociation - dissociates the ions using neutral gases and then recombines them by capturing low energy electrons using multiply charged ions (used for peptides and proteins)
Electron - Transfer Dissociation - multiply charged gaseous molecules transfer electrons. ion-ion reaction and from a CI source
QIT
Disdvantages:
product ion scan only
Low mass cut off
spectra more complex
Advantages:
Simple to operate
MSn is possible
full scan sensitivity
higher m/z than QQQ
QQQ
Advantages:
Simple to operate
sensitive to SIM and SRM
No loss mass cut off
Disadvantages:
MS/MS only
Low resolving power
Acquisition is 1 spectrum/second
Maximum m/z range is 4000
Gas Phase:
Electron Impact Ionisation - Energetic Electrons
Chemical Ionisation - ReagentGaseous Ions (NH3, CH4 or C4H10)
Photo-ionisation
Liquid Ionisation:
Atmospheric Pressure Ionisation
Electrospray Ionisation
Atmospheric Pressure Chemical Ionisation
Atmospheric Pressure Photo-ionisation
Ionisation occurs outside the vacuum region
All soft ionisation techniques
For polar, low volatility (high molecular mass) and thermolabile (changes based on heat) compounds
Mechanism of ionisation has the same principles for all of them which is desolvation and ion transmission
Soft Ionisation - leads to fewer fragments, lower energy and gives the high molecular ion
Hard Ionisation - will have higher energy, will give more fragments, and will give a low molecular ion, produces predictable ions
The space between the sample and the grid is the accelerationregion the flight tube is the Ldriftpath.
At the grid, the weird ions are neutralised - similar to ion optics
Linear TOF Analyser:
Upper m/z range has not limits
Has higher transmission efficiencies therefore -
Very sensitive
Fast (covers a wide range in microseconds)
gives 1000 of spectra
However
if an ion packet energy is normally distributed then there will be less kinetic energy and therefore issues with resolution resulting in broader peaks
Mass Resolution of TOF:
TOF has poor mass resolution ---> due to distribution of flight times among the ions with the same m/z ---> space it takes up, the time it takes and the speed it then goes at leads to peak broadening.
TOF Extraction of Ions:
Continuous Extraction - ions have the same m/z but different velocities (still give broad peaks). Acceleration region starts at 20kV
Delayed Pulsed Extraction - acceleration region starts at 20kV then half way between the sample and the grid there is another 20kV region acting in a pulse way - helps push the smaller ions further
Ion Extraction - acceleration region starts at 26kV then halfway before the grid there is another 20kV, timedelay between the two voltages - helps peaks get wider and narrower
Reflectron TOF:
cheaper but only works to a certain concentration
The flight tube has a bend to it, and where it bends there are ringelectrodes that have increased voltages that causes the largerions to repel
the increased voltage will allow the smaller ions to stay on the flight tube but it also increases the flightpath length as it creates a broader bend
Orthogonal Acceleration TOF
beam goes down the flight tube and hits reflectron lenses that reflect the beamback to hit the detector near where the beam enters the flight tube.
Tandem TOF
can be done with quadrupole
used for Post Source Decay and generates metastable ions
Electron Ionisation:
Hard ionisation technique
Vaporisation of the sample before ionisation
Only for volatile and thermostable compounds
enhance volatility and stability for polar analytes by derivatisation
used in GC-MS more than LC
Chemical Ionisation:
Soft ionisation technique
CI source is a reagentgas inside a modified EI source
no extensive fragmentation
high sensitivity and selectivity
only for volatile and thermostable compounds
ESI Flow Rate:
Depending on the flow rate three types of ESI can be done
Conventional ESI - high flow rate, for larger molecules that have a difference of 20 mm compared to the smallest molecules
Micro ESI - flow rate less than 1 ul/min for molecules that have a difference of 2 mm
Nan ESI - flow rate less than 0.5 ul./min and a molecule size difference of 1-2 mm
This is down to the source and the sensitivity
ESI:
Advantages:
soft ionisation
generates ions from liquid phase
efficient ion production
suitable for a wide range of analytes
forms multiply charged ions for the bigger molecules
Disadvantages:
sensitive to matrix effects
Adduct formation which decreases signal and complicates results