Collimation to a fan x-ray with associated scatter radiation rejection and improvement in image contrast
The x-ray beam is collimated to a fan by pre-patient collimators
Post-patient image-forming x-rays likewise are collimated to a fan that corresponds to the detector array (grid) - a scintillation phosphor, usually Sodium Iodide (NaI) OR Cesium Iodide (CsI) - and is married to a linear array of CCDs (charge-coupled device) through a fiber optic light path
Scanned Projection Radiography (SPR) was not very successful because chest anatomy has high subject contrast, so scatter radiation rejection is not all that important
A successful approach to DR where the scintillation light from a CsI phosphor is efficiently transmitted through fiber optic bundles to the CCD array, resulting in high x-ray capture efficiency and good spatial resolution
The assembly of multiple CCDs for the purpose of viewing an area x-ray beam presents the challenge to create a seamless image at the edge of each CCD, which is accomplished by interpolation of pixel values at each tile interface