CC LEC M3 U1

    Cards (55)

    • Module Objectives:
      • Differentiate the methods of instrumentation as to their principle of operation, components, advantages, limitations, and application in the measurement of different analytes
      • Practice appropriate preventive maintenance on different instruments used in clinical chemistry
    • Module Self-Monitoring Form activities include reading module introduction, contents, and objectives, engaging, exploring, explaining, elaborating, and evaluating different topics like spectrometry, luminescence, electro-analytical methods, chromatography, automation, and point of care testing
    • Unit 1: Spectrometry:
      • Engage, explore, explain, elaborate, and evaluate the topic
    • Unit 2: Luminescence:
      • Engage, explore, explain, elaborate, and evaluate the topic
    • Unit 3: Electro-analytical Methods:
      • Engage, explore, explain, elaborate, and evaluate the topic
    • Unit 4: Chromatography:
      • Engage, explore, explain, elaborate, and evaluate the topic
    • Unit 5: Automation:
      • Engage, explore, explain, elaborate, and evaluate the topic
    • Unit 6: Point of Care Testing:
      • Engage, explore, explain, elaborate, and evaluate the topic
    • Analytic techniques and instrumentation provide the foundation for all measurements made in a modern clinical chemistry laboratory
    • The majority of techniques in analytic chemistry fall into four basic disciplines: spectrometry, luminescence, electroanalytic methods, and chromatography
    • Spectrophotometry/Photometry/Colorimetry involves the measurement of radiant energy emitted, transmitted, absorbed, or reflected under controlled conditions
    • Spectrophotometry uses a monochromator to select the desired region of the spectrum for measurement, with slits isolating a narrow beam of light to improve its chromatic purity
    • Beer's Law explains the relationship between absorbance at a given wavelength and concentration, expressed as A = abc, where A is absorbance, a is absorptivity, b is light path, and c is concentration
    • Percent transmittance is the ratio of radiant energy transmitted divided by the radiant energy incident on the sample, with 0% T indicating all light absorbed and 100% T when no light is absorbed
    • Absorbance is defined as -log(I/Iâ‚€) = log(100%) - log % T = 2 - log % T
    • According to Beer’s law, absorbance is directly proportional to concentration
    • A spectrophotometer is used to measure the light transmitted by a solution to determine the concentration of the light-absorbing substance in the solution
    • Components of a Spectrophotometer:
      • Light Source:
      • Incandescent Lamps:
      • Tungsten/Tungsten Iodide Lamp: emits light in the visible portion of the spectrum, with most emitted in the IR region
      • Hydrogen and deuterium lamps provide continuous spectra in the UV region
      • Low Pressure Mercury Arc Lamp emits a discontinuous or sharp line spectrum with both UV and visible lines
      • Hollow cathode lamp has a very-narrow-wavelength intense source
      • Lasers Sources: used as a light source in spectrophotometers
      • LED (Light Emitting Diodes) made up of semiconductor (aluminum gallium arsenide) and insulator
      • Device for Spectral Isolation achieved by monochromators
    • Filters used as monochromators:
      • Glass filter: simple filter made of colored glass
      • Sharp-cutoff filter: eliminates light below a given wavelength
      • Narrow-bandpass filters: constructed by combining sharp cutoff filters or regular filters
    • Prisms and Gratings:
      • Prism: separates white light into a continuous spectrum through refraction
      • Diffraction grating: prepared by depositing a thin layer of aluminum-copper alloy on a glass plate, then making small parallel grooves in the metal coating
    • Sample Cell:
      • Also known as a cuvet/cuvette or absorption cells
      • Used to hold a liquid sample to be analyzed in the light path of a spectrometer
      • Cuvets may be round, square, or rectangular, with advantages in accuracy and less error from the lens effect
    • Quartz cells are used for readings below 340 nm, while plastic cells are designed for disposable, single-use applications
    • Cuvets must be clean and optically clear; scratched optical surfaces scatter light and should be discarded
    • Cuvets used in the UV region should be free from invisible scratches, fingerprints, or residual traces of previously measured substances which may significantly affect absorbance readings
    • Photodetectors convert transmitted radiant energy into electrical energy proportional to the number of photons striking its photosensitive surface
    • A photomultiplier (PM) tube is commonly used for measuring light intensity in the UV and visible regions of the spectrum; it has extremely rapid response times, is very sensitive, and has adequate sensitivity over a wide wavelength range
    • Solid-state detectors like photodiodes are capable of measuring light at a multitude of wavelengths; when consisting of a two-dimensional array of diodes, each responding to a specific wavelength, it is known as a photodiode array
    • Charge-coupled detectors are multi-channel devices with good dynamic ranges and signal to noise ratios superior to those of PM tubes; they are used for molecular fluorescence measurement of very low concentrations of fluorescent molecules
    • Photocells, also known as barrier-layer cells, are used in older instruments; they are inexpensive and durable but temperature-sensitive and non-linear at very low and very high levels of illumination
    • Phototubes are similar to photocells but require an outside voltage to operate and are more sensitive; they contain an anode and cathode in a glass case
    • Readout devices display electrical energy from a detector on a meter or readout system; they can be direct reading or null point systems, with digital readout devices providing visual numerical displays of absorbance or converted values of concentration
    • A single beam spectrophotometer is composed of the basic components of a spectrophotometer, deriving a single measurement where the concentration of a solution can be determined
    • A double beam spectrophotometer splits light into two beams before reaching the sample; the measurement displayed comes from the ratio of the two beam intensities
    • Flame photometry is based on the principle that excited atoms, when returning to the ground state, emit light energy characteristic of the atomic species
    • Flame photometers are no longer routinely used in clinical chemistry laboratories due to the development of ion selective electrodes for analytes like Na, K, or Li
    • Flame photometry can determine only the concentrations of pure metals; it involves an aspirator, atomizer, flame, lenses, monochromator, and a detector
    • The FEP monochromators used for common analytes are: Sodium (589nm) emitting a yellow flame, potassium (767nm) emitting a violet flame, and lithium (671nm) emitting a red flame
    • Advantages of flame emission photometry include an accuracy and precision of +/- 1-5% in aqueous solution, suitability for many metallic ions, and being a fast, simple analytical method for determining trace metal ions in solution with relatively few interferences from other elements
    • Flame Atomic Emission Spectrophotometry (FEP) is used in routine clinical labs primarily for sodium and potassium measurements
    • Lithium levels in the serum are determined by FEP for manic-depression patients on lithium therapy
    See similar decks