Differentiate the methods of instrumentation as to their principle of operation, components, advantages, limitations, and application in the measurement of different analytes
Practice appropriate preventive maintenance on different instruments used in clinical chemistry
Module Self-Monitoring Form activities include reading module introduction, contents, and objectives, engaging, exploring, explaining, elaborating, and evaluating different topics like spectrometry, luminescence, electro-analytical methods, chromatography, automation, and point of care testing
The majority of techniques in analytic chemistry fall into four basic disciplines: spectrometry, luminescence, electroanalytic methods, and chromatography
Spectrophotometry/Photometry/Colorimetry involves the measurement of radiant energy emitted, transmitted, absorbed, or reflected under controlled conditions
Spectrophotometry uses a monochromator to select the desired region of the spectrum for measurement, with slits isolating a narrow beam of light to improve its chromatic purity
Beer's Law explains the relationship between absorbance at a given wavelength and concentration, expressed as A = abc, where A is absorbance, a is absorptivity, b is light path, and c is concentration
Percent transmittance is the ratio of radiant energy transmitted divided by the radiant energy incident on the sample, with 0% T indicating all light absorbed and 100% T when no light is absorbed
A spectrophotometer is used to measure the light transmitted by a solution to determine the concentration of the light-absorbing substance in the solution
Prism: separates white light into a continuous spectrum through refraction
Diffraction grating: prepared by depositing a thin layer of aluminum-copper alloy on a glass plate, then making small parallel grooves in the metal coating
Cuvets used in the UV region should be free from invisible scratches, fingerprints, or residual traces of previously measured substances which may significantly affect absorbance readings
A photomultiplier (PM) tube is commonly used for measuring light intensity in the UV and visible regions of the spectrum; it has extremely rapid response times, is very sensitive, and has adequate sensitivity over a wide wavelength range
Solid-state detectors like photodiodes are capable of measuring light at a multitude of wavelengths; when consisting of a two-dimensional array of diodes, each responding to a specific wavelength, it is known as a photodiode array
Charge-coupled detectors are multi-channel devices with good dynamic ranges and signal to noise ratios superior to those of PM tubes; they are used for molecular fluorescence measurement of very low concentrations of fluorescent molecules
Photocells, also known as barrier-layer cells, are used in older instruments; they are inexpensive and durable but temperature-sensitive and non-linear at very low and very high levels of illumination
Phototubes are similar to photocells but require an outside voltage to operate and are more sensitive; they contain an anode and cathode in a glass case
Readout devices display electrical energy from a detector on a meter or readout system; they can be direct reading or null point systems, with digital readout devices providing visual numerical displays of absorbance or converted values of concentration
A single beam spectrophotometer is composed of the basic components of a spectrophotometer, deriving a single measurement where the concentration of a solution can be determined
A double beam spectrophotometer splits light into two beams before reaching the sample; the measurement displayed comes from the ratio of the two beam intensities
Flame photometry is based on the principle that excited atoms, when returning to the ground state, emit light energy characteristic of the atomic species
Flame photometers are no longer routinely used in clinical chemistry laboratories due to the development of ion selective electrodes for analytes like Na, K, or Li
Flame photometry can determine only the concentrations of pure metals; it involves an aspirator, atomizer, flame, lenses, monochromator, and a detector
The FEP monochromators used for common analytes are: Sodium (589nm) emitting a yellow flame, potassium (767nm) emitting a violet flame, and lithium (671nm) emitting a red flame
Advantages of flame emission photometry include an accuracy and precision of +/- 1-5% in aqueous solution, suitability for many metallic ions, and being a fast, simple analytical method for determining trace metal ions in solution with relatively few interferences from other elements