— a device that acts as the primary control mechanism for the entire fluoroscope.
— It is through the X-ray generator that current is allowed to flow into the X-ray tube.
The basic function of adjusting the voltage differential and current of the X-ray tube are controlled automatically to maintain optimal contrast and brightness
— X-rays may be generated in either a continuous or a pulsed mode.
— Automatic brightness control is a standard feature of the majority of modern fluoroscopes. Through this system mA and kVp are constantly monitored and adjusted to optimize the image.
X-RAY TUBE
— Functions as a specific energy converter,
— receiving electrical energy and converting it into two other forms of energy:
x-radiation (1%) and
heat (99%).
— Heat is considered the undesirable product of this conversion process; therefore x-radiation is created by
— taking the energy from the electrons and converting it into photons.
— ELECTRONS TO PHOTONS:
— This very specific energy conversion takes place in the x-ray tube.
CATHODE (━)
· negativeside of the x-raytube;
· it has two primary parts, a filament and a focusingcup.
· The twofilamentssupply separate electron beams to produce two focalspots.
Filament
· Is a coil of wire similar to that in a kitchen toaster, but it is much smaller.
· Approximately 2 mm in diameter and 1 or 2 cm long.
· Emits electrons when it is heated.
When the current through the filament is sufficiently high, the outer-shell electrons of the filament atoms are “boiled off” and ejected from the filament. This phenomenon is known as THERMIONIC EMISSION.
Focusing Cup
· The filament is embedded in a metal shroud called the focusing cup
· Because all of the electrons accelerated from cathode to anode are electrically negative, the electron beam tends to spread out owing to electrostatic repulsion.
· Some electrons can even miss the anode completely.
ANODE (+)
● Is the positive side of the x-ray tube
● It conducts electricity
● Radiates heat
● Contains the TARGET
TARGET - o Area of the anode struck by the electrons from the cathode
FUNCTIONS
o Conducts electrons from the cathode
o Mechanical support for the target
o Thermal dissipater
IMAGE INTENSIFIER
· Complex electronic device that receives the image-forming x-ray beam and converts it into a visible-light image of high intensity
Input phosphor = Cesium iodide (CsI)
Output phosphor
● Usually made of zinc cadmium sulfide.
● The site where electrons interact and produce light.
Electron optics
● Maintaining proper electron travel
● Electrostatic focusing lenses
● Directs the path of electrons towards the anode
INPUT PHOSPHOR-X-ray energy is converted into visible light
Similar to the effect of radiographic intensifying screens
Photocathode
● Usually composed of cesium and antimony compounds that
● Respond to stimulation of input phosphor light by the emission of electrons
● The photocathode emits electrons when illuminated by the input phosphor.
● Photoemission - electron emission that follows light stimulation.
The lenses and the mirror in the TV lens system must remain precisely adjusted to avoid a blurred image
The objective lens accepts light from the output phosphor and converts it into a parallel beam
An image recorded on film is interrupted by a beam-splitting mirror
The beam-splitting mirror transmits only a portion of the beam to the television camera
The beam-splitting mirror is retracted from the beam when a film camera is not in use
Both the television camera and the film camera are coupled to lenses that focus the parallel light beam onto the film and target of the respective cameras
TV camera IS Used to electronically convert the visible image on the output phosphor of the image intensifier.
The television camera consists of cylindrical housing, approximately 15 mm in diameter by 25 cm in length, that contains the heart of the television camera tube.
TV CAMERA - It also contains electromagnetic coils that are used to properly steer the electron beam inside the tube. A number of such television camera tubes are available for television fluoroscopy, but the vidicon and its modified version, the Plumbicon, are used most often.
The video signal received by the television picture tube is modulated, that is, its magnitude is directly proportional to the light intensity received by the television camera tube.
Different from the television camera tube, the electron beam of the television picture tube varies in intensity according to the modulation of the video signal
· The intensity of the electron beam is modulated by a control grid, which is attached to the electron gun.
This electron beam is focused onto the output fluorescent screen by the external coils
Photographic Subtraction Technique
⚫ Introduced by Ziedses des Plantes, a technique by which bone structure images are subtracted, or cancelled out, from a film of the bones plus opacified vessels, leaving an unobscured image of the vessels
⚫ Zero Film or base or control film – shows bony structures only, no patient motion must be present in this film. Zero film is exposed just before contrast medium is injected into vessels
⚫ Series film – contains image post-injection of contrast material
⚫ Registration - matching of one image over another so that bony landmarks are precisely superimposed
ray generator is the primary control mechanism for the entire fluoroscope
Current flows into the X-ray tube through the X-ray generator
ray generator automatically adjusts the voltage differential and current of the X-ray tube to maintain optimal contrast and brightness
High-frequency generators provide superior exposure reproducibility, compact size, lowest purchase price, and lowest repair costs
rays can be generated in either continuous or pulsed mode
Automatic brightness control is a standard feature in modern fluoroscopes